There a few issues with this software.
1) I can save my setting to a file but when I recall the file, the settings revert back to the defaults.
2) There's no refresh button on the waterfall screen It would be nice if one could clear the slate and start with a new waterfall.
3) On the low end of the spectrum, in the audio range, the calibration is off.
4) The resolution bandwidth in inadequate in the audio range.
1) I can save my setting to a file but when I recall the file, the settings revert back to the defaults.
2) There's no refresh button on the waterfall screen It would be nice if one could clear the slate and start with a new waterfall.
3) On the low end of the spectrum, in the audio range, the calibration is off.
4) The resolution bandwidth in inadequate in the audio range.
> There a few issues with this software.
>
> 1) I can save my setting to a file but when I recall the file, the settings revert back to the defaults.
>
I have seen this, too. I'll pass it along.
> 2) There's no refresh button on the waterfall screen It would be nice if one could clear the slate and start with a new waterfall.
>
I agree 100%. I'll pass it along.
> 3) On the low end of the spectrum, in the audio range, the calibration is off.
>
> 4) The resolution bandwidth in inadequate in the audio range.
This is probably the wrong tool for working in the audio range. It was designed for use in RF measurements. Resolution BW is a function of trace depth, sample rate, and window type. For a number of reasons, mostly related to SW and dataset size, the minimum sample rate on the scope when using ASV appears to be 20M, which limits the minimum RBW to 36Hz.