as i sow in the specification of nfa n8975a its could be possible to measure nf of a device that its noise figure greater than the ENR of the noise source. as i know the formula of noise figure is NF(db) = ENR (db) -10log (y-1). as i understand the ENR OF THE NOISE SOURCE MUST BE GREATER THAN THE NOISE FIGURE OF THE DEVICE. CAN MODERATOR EXPLAIN ME THIS MATTER.
THANK YOU
THANK YOU
When the noise figure of a device is higher than the ENR of the noise source, the device noise tends to mask the noise source output. In cases such as this, the Y-factor will be very close to 1. Accurate measurements of small ratios can be difficult. Generally the Y-factor method is not used when the noise figure is more than 10 dB above the ENR of the noise source. For this measurement challenge there are two suggested methods:
1) The signal generator twice-power method
2) Direct noise measurement method
See application note 57-1 from Agilent Technologies.
Regards -