How to save the "Corrected Flatness" data for the frequency response adjustment in the ESA-L1500A if not using the adjustment software? Does anybody know the prrocedure for that?
The ESA-L1500A could only be adjusted via Agilent software. The frequency response adjustment test came up with the corrections factors and then stored those corrections to analyzer memory. I haven’t heard of a manual procedure. In fact, a big majority of test equipment today cannot be adjusted manually; it requires the appropriate software.
The ESA-L1500A could only be adjusted via Agilent software. The frequency response adjustment test came up with the corrections factors and then stored those corrections to analyzer memory. I haven’t heard of a manual procedure. In fact, a big majority of test equipment today cannot be adjusted manually; it requires the appropriate software.
Regards -