Hi there,
I'm trying to connect to an MSO6032A scope using Intuilink Data Capture (V3.51) over LAN.
I have installed I/O Libraries V15.0.11221.0 and this appraently connects to the scope with no problems.
However when I start Intuilink it can't find the instrument - when I select "Agilent 6000 Series" from the list the i/o address is fixed on GPIB, and when I click on Find Instrument it searches busses and comes up with a dialog only containing COM1.
Both utilities were downloaded from the Agilent website today...
Any ideas? (I can access the scope via web browser, so clearly have connection to it.)
Thanks,
Oliver.
I'm trying to connect to an MSO6032A scope using Intuilink Data Capture (V3.51) over LAN.
I have installed I/O Libraries V15.0.11221.0 and this appraently connects to the scope with no problems.
However when I start Intuilink it can't find the instrument - when I select "Agilent 6000 Series" from the list the i/o address is fixed on GPIB, and when I click on Find Instrument it searches busses and comes up with a dialog only containing COM1.
Both utilities were downloaded from the Agilent website today...
Any ideas? (I can access the scope via web browser, so clearly have connection to it.)
Thanks,
Oliver.
Also, did you specify it do an *IDN? or did you specify WEB connection?
You can turn on the IO monitor in the IO libraries and then
start Intuilink. Look to see if it tries to interrogate your scope
in the startup.