I am trying to get to the attenuator on the acquisition board. The Agilent Service Manual says to 1st remove the acquisition board. I go to the acquisition board removal and the procedure says to remove the Quick-Probe subpanel assy. Now the service manual calls the quick-probe the Auto-Probe . But anyway the procedure says to use a small screw driver or slender pointed object through the access hole between attenuator 2 & 3.and push the Auto-Probe assy away from the panel assy. Now I tried this and the probe assy does not want to come away from the front panel assy. Is there something else that I either should be removing before the auto probe assy will pull away from the front panel assy. Is removing the Auto-Probe assy really necessary to remove the acquisition board?
I am trying to get to the attenuator on the acquisition board. The Agilent Service Manual says to 1st remove the acquisition board. I go to the acquisition board removal and the procedure says to remove the Quick-Probe subpanel assy. Now the service manual calls the quick-probe the Auto-Probe . But anyway the procedure says to use a small screw driver or slender pointed object through the access hole between attenuator 2 & 3.and push the Auto-Probe assy away from the panel assy. Now I tried this and the probe assy does not want to come away from the front panel assy. Is there something else that I either should be removing before the auto probe assy will pull away from the front panel assy. Is removing the Auto-Probe assy really necessary to remove the acquisition board?
You need to remove the nuts from the inputs on front panel before panel will come out. nuts are under the snap-in faceplate at the bottom right (one with channels are sticking out). You have to push out the panel from the behind bottom of the front hey board and display assembly.
mickey