Hi
RMS phase jitter opt 226 ver B.01.72 doesn't seem to work the way the instructions say (5988-3698). We need to enter the start and stop frequencies for the integration, and the manual says you have to go [Marker] {Normal}, but as soon as I do this I can't enter frequency. I have found by trial and error you have to go
[Marker] then enter the start frequency, then
{rms noise} then {degs/rads/jitter} then
[Marker] and enter the stop frequency
Anyone else seen this?
Rob Bristow
RMS phase jitter opt 226 ver B.01.72 doesn't seem to work the way the instructions say (5988-3698). We need to enter the start and stop frequencies for the integration, and the manual says you have to go [Marker] {Normal}, but as soon as I do this I can't enter frequency. I have found by trial and error you have to go
[Marker] then enter the start frequency, then
{rms noise} then {degs/rads/jitter} then
[Marker] and enter the stop frequency
Anyone else seen this?
Rob Bristow
I just noticed this older posting and I took a look for 5988-3698, but I can't find this part number anywhere. Can you please confirm the part number?
Also, here is a link to the most recent version of the PSA Option 266 manual, part number E4440-90328.
http://www.home.agilent.com/agilent/red ... 11143.0.00
Would you mind taking a look to see if this most recent version of the manual has this same issue?