Greeting,
I encountered a problem on moving the marker from marker delta value via sending commands through GPIB with labview. There is a series of commands but the PSA screen was still remaining unchanged with marker still.
The setting that I set to PSA 4445A Spectrum is:
- Set a test frequency example 136MHz with span to zero;
- Resolution BW to 100 KHz
- Set a Video trigger line to capture the signal rise time
- Then max peak the signal and delta to 0.
After that, I am trying to move the marker to -6dB below the delta peak but the PSA didn't perform as the command.
The command that I sent to PSA are:
1. :CALC:MARK1:MAX
2. :CALC:MARK1:MODE DELT
3. :CALC:PST:MARK1:X:POS 10
However, the last command is not working and the PSA didn't return any error.
Please help to resolve my problem.
Thanks and Regards.
1. :CALC:MARK1:MAX
Thanks and Regards.
The :CALC:PST:MARK1:X:POS 10 SCPI command is only used in the SA measurement mode Power Stat CCDF. There is no specific command to move the delta marker to a specific y-axis offset. You can specify the x-axis position of the delta marker based on its display point.
:CALCulate:MARKer[1]|2|3|4:X:STARt <param>
:CALCulate:MARKer[1]|2|3|4:X:STOP <param>
Maybe you should try exporting the trace data to your PC and then finding the required dB offset.