We are using the Measuring Receiver system (Option 233 PSA) for Signal Generator calibration.
I have a few doubts in the Tuned RF Level Measurement,
I could understand that the Range 2 Switch Level and Range 3 Switch level are the input level at which the measuring system changes the range of measurement. ( we make the Range switching Auto)
But,
What is the use of Cal Factor?
Does the cal factor value has any affect in the Power measurement displayed?
Do we need to use that value in calculating the original value of the Power?
Thanks for your help in Advance,
Pandiarajan R
But,
Pandiarajan R
> What is the use of Cal Factor? Does the cal factor value has
The Cal Factors are the calibration factors for your power sensor module. These factors are created for each sensor and they must be used when you zero and cal your sensor before making a measurement order to attain the specified measurement accuracy.
The Cal Factors come on a floppy disk when you send your sensor in for periodic calibration. The process for loading the cal factors is on page 52 of the Measuring Receiver manual.
http://cp.literature.agilent.com/litweb ... -90003.pdf
Best regards,