Hi folks,
Trying to acquire data from the 3024A. I'm using the agilent-provided "Agilent 2000 3000 X-Series Acquire Waveform.vi" example, without any changes. I can acquire channel 1 just fine. If I change to any other channel the VISA read in the Read Waveforms(Single).vi fails with a "Query Unterminated" and "Settings Conflict" error. I've probed the VISA strings used to set up the transfer and there are identical except for channel number.
Any ideas?
Thanks,
Stephenf555
LabView 10.0f2
Most recent drivers and software for Oscilloscope
Trying to acquire data from the 3024A. I'm using the agilent-provided "Agilent 2000 3000 X-Series Acquire Waveform.vi" example, without any changes. I can acquire channel 1 just fine. If I change to any other channel the VISA read in the Read Waveforms(Single).vi fails with a "Query Unterminated" and "Settings Conflict" error. I've probed the VISA strings used to set up the transfer and there are identical except for channel number.
Any ideas?
Thanks,
Stephenf555
LabView 10.0f2
Most recent drivers and software for Oscilloscope
After looking around it seems your problem has been resolved by National Instruments. I tested their new VI's for the 2000x and 3000x and they all seem to be working properly. The link for the driver download is below:
http://sine.ni.com/apps/utf8/niid_web_d ... 21287E6A9E
If you run in to any more problems please let us know.