I won't have the analyzer in front of me again until Monday and couldn't find anything in the .pdf.
I'd like to know if the 4407B can do an offset stimulus response measurement (i.e. measure 500MHz and stimulate 600MHz). I know there are PSA models that can do this, but wasn't sure if the 4407B can do it.
Any help here would be greatly appreciated.
I'd like to know if the 4407B can do an offset stimulus response measurement (i.e. measure 500MHz and stimulate 600MHz). I know there are PSA models that can do this, but wasn't sure if the 4407B can do it.
Any help here would be greatly appreciated.
Unfortunatly there is no way to set the ESA tracking generator frequency to be offset from the SA frequency.
You are right about PSA, with an external source and source control option it has a lot of flexibility with making these types of measurements.