I set the wavegenerator on my DSOX 2002A to
SIN, 10 Hz 300mVpp
I measure with the scope
Measurementfunction tells me Pk-Pk(1) 593 mV
Now i change the generator setting to Highlevel/Lowlevel
High is set to 150 mV
Low is set to -150 mV
Measurementfunction tells me Pk-Pk(1) 593 mV
I would expect 300 mV now.
Is is a bug or my mistake? Where should i report the bug, if it is a bug?
Kind regards
SIN, 10 Hz 300mVpp
I measure with the scope
Measurementfunction tells me Pk-Pk(1) 593 mV
Now i change the generator setting to Highlevel/Lowlevel
High is set to 150 mV
Low is set to -150 mV
Measurementfunction tells me Pk-Pk(1) 593 mV
I would expect 300 mV now.
Is is a bug or my mistake? Where should i report the bug, if it is a bug?
Kind regards
If it is set to 50 Ohms, and you are connecting to a 1M input on the scope, then you should see about twice what you expect, which is what I think is happening. If you change either impedance, so that they match, you should see the correct values.