I am new to DSOX2024A built in function generator. I want to connect a load to the wavegen output. I set the wavegen as PULSE wave-250Hz-5Vpp-width(2ms). when I check this output with the DSO, I could see the perfect PULSE wave-250Hz-5Vpp-width(2ms). But when I try to connect the load to the wavegen output. I couldn't see the pulse wave in the DSO I could see only the fully OFF signal(0V), I am using the Agilent N2863B 300 MHz PASSIVE PROBE.
Please suggest me the best way to connect the function generator to the load.
Please suggest me the best way to connect the function generator to the load.
Output impedance 1M, no load - Full voltage
Output impedance 1M, 50 Ohm load - Very small voltage
Output impedance 50 Ohm, no load - Double voltage
Output impedance 50 Ohm, 50Ohm load - Full voltage
Al