Are there any known issues with option 122 being lost after a firmware upgrade on PSA (E4440A)?
At the point in the update process when firmware for option 122 is loaded, the PSA screen indicates “Waiting for WB DIF to initialize†for a long time and eventually times out. Then the update program reports “Error +1000: “Failed to write to staging file†and then goes on to load the firmware for the next option.
After the update program is finished and the PSA is rebooted, it reports “WBIF has incompatible revisionâ€. The PSA appears to work fine in spectrum analyzer mode; however, if I place the instrument in basic mode and try to set it to wideband mode, the option does not appear. If I try to list the the installed options, option 122 no longer appears, however the license key is listed when I check for it.
At the point in the update process when firmware for option 122 is loaded, the PSA screen indicates “Waiting for WB DIF to initialize†for a long time and eventually times out. Then the update program reports “Error +1000: “Failed to write to staging file†and then goes on to load the firmware for the next option.
After the update program is finished and the PSA is rebooted, it reports “WBIF has incompatible revisionâ€. The PSA appears to work fine in spectrum analyzer mode; however, if I place the instrument in basic mode and try to set it to wideband mode, the option does not appear. If I try to list the the installed options, option 122 no longer appears, however the license key is listed when I check for it.
The fix is to change the PSA IP address (192.168.100.2 for PC and 192.168.100.3 for PSA work well) and then reload firmware. The firmware upgrade will install the core firmware, then install the personalities and WBIF firmware.
You must then cycle power after the personalities have been loaded for the WBIF firmware to reload into the WBDIF before it will work again.