Hi,
What is the max trigger rate when acquisition is in averaging mode on the infiniium series?
I test the following:
I use a burst of 16 pulses and set the number of averages to 16 on the DSO80204. If the separation between the pulses is less than 200ms then trigger events are lost and the averaging is not complete after a single burst. The 200ms is independent of the number of samples in the waveform.
Can the trigger rate be reduced?
Thanks
Carl
What is the max trigger rate when acquisition is in averaging mode on the infiniium series?
I test the following:
I use a burst of 16 pulses and set the number of averages to 16 on the DSO80204. If the separation between the pulses is less than 200ms then trigger events are lost and the averaging is not complete after a single burst. The 200ms is independent of the number of samples in the waveform.
Can the trigger rate be reduced?
Thanks
Carl
You might try using High Resolution mode instead of averaging. This is done in hard ware and is not dependent upon a trigger event to happen for the next set of data to be averaged with the previous
Some other things to would be to decrease the memory points or increase the sampling rate.
Mark