Model : N9938A (Spectrum Analyzer)
Refer to Rec. ITU-R SM.1268-1
Purposer : For 104.5 MHz, I want to set a Spectrum Analyzer to
- span: 340 kHz;
- sweeptime: 340 ms (1 ms/kHz)
Problem :I can configure span = 340 kHz but, I cannot configure the sweep time value to 340 ms. (exactly value)
Is N9938A supported to configure the sweeptime to be 340 ms in the continuous mode?
If support, how is a step to manually set sweeptime = 340 ms.
On the N9938A, you can set the sweep acquisition parameter, which provides results similar to sweep-time. Setting the sweep acquisition to 400 will results in approximately 340 ms of sweep time.
[Sweep] {Sweep Acquisition} 400
Afsi Moaveni