Regarding my E5062A I see that in the pull-down menu for cal kits, when the 85032 kit is opened up and examined I only see a total of (4) CLS definitions yet there are (6) components in that cal kit, OPEN(m), OPEN(f), SHORT(m), SHORT(f), THRU, and LOAD. Should there not be a CLS associated with each element in a cal kit? I wish the manual talked a little bit about the function of CLASS (CLS) and how they are used.
You didn't mention which 85032 kit you inspected on your E5061A. From the description, I'm assuming it was the 85032F because you said there wasn't a standard for all 6 devices in the kit. From the VNA support web site, you can see the most current definitions for all current cal kits. Keysight Cal Kit Page If you click on the ENA cal kit for 85032F, you will see that the definitions for the opens and shorts are the same for male and female. So the ENA firmware probably supplies one set of coefficients for shorts and opens and refers both sexes to the same set. I don't have an ENA in front of me right now, but I'm pretty sure that's what you're seeing. Look at the 85032B and you will find that male and female are different there.
DH