my 36313A can't be used for both voltage and current in STEP mode,neither the BOST and EBOT in output list.
the maunual described not details.Can you tell me how to use them?
my 36313A can't be used for both voltage and current in STEP mode,neither the BOST and EBOT in output list.
the maunual described not details.Can you tell me how to use them?
I have read the manual over and over again.but I still can not use these function freely.
1.the “BOST” and "EOST" in output list,I can't get a output signal althougth I set the IO function to "trigger out".
2.the output mode "step", can't be used with "fixed",when I set the mode as this,turn back to see the mode,it turned to all the "list".above all,it looks like can only work in "list"and "fix" mode.once you set it to others,it will all back to "list".where I did wrong?can you help me?
thank you in advance.
Hi
I do not know this device, it is difficult to find the solution directly, but I have already used the LIST and STEP mode on other Keysight devices, and operation seems similar.
STEP mode will have a LIST that will advance one step at a time for each received TRIG signal. BOST and EOST can be configured like the example below. Voltage and current must be have the same setup.
Mode LIST and STEP don't works together, it's normal two differents functions.
But for STEP and FIXED, I'm not sure. If so, then, I guess you should not write anything in the boxes of the list, except box 0
I do not know if these few remarks can help you.
If you find the solution, I would be curious to know it.
Yves
hi,Yves.Sincerely thanks for your so detailed help.but my equipment still have a falult response.Please look at my attached video.
Once I set the volt & current mode to all "Step",press the "Run/Stop",return to "Properties",you can see all the mode turn to "List"!!This is annoying.what I can do?please help.Or if it is possible you have a operate video of teach how to use this step for our users?
another question I forgot to ask is,the "Digital IO" and "Digital In",how I can use them? the IO interface have no response either.I really don't understand this IO function,I bought this model,but these functions i have not met before.
thank you for your help.
sincerely
Hi,
Does the device execute the requests even if it returned to LIST mode. If so, then I wonder if when the execution list is finished, if it does not automatically return to LIST. In the repeat count, there is only one try, try a higher count and see the behavior. Make even a test with the CONTINUOUS mode.
For the 3 available pins of control I/O, you must configure them before using them.
Example below:
For TRIG OUT
- Press utilities
- I/O config
- Digital IO
- Choose pin number 1 (1,2 or 3)
- Function:Trigger OUT
- Polarity: Positive
Now If you use BOST and EOST, you will have pulse on the pin #1
For example to have Trigger IN, just change slection "Trigger out" to "Trigger in"
Inhibit in is use to control output state
Digital I/O is a basic input-output control.
Yves
Hi, in the user manual user manual E36300 Series beginning page 77. Have you followed all the steps? What does not work exactly?
Yves