I have several 34450A multimeters and I have a problem with restoring settings after turning off.
I can save and recall settings, but the multimeter is not loading settings after rebooting from memory "Last".
I am trying to set that after startup the default measurement would be, for example, resistance measurement with the resolution "Medium", but the factory settings are always loaded.
Is this a bug in the software, should it be this way?
Regards
Miro112
Hi, I verified this setup with my 34410A: difference, I have 5 registers instead of 6, and the name of register 0 is POWER_DOWN instead of LAST for the 34450A.
You have 6 registers for save setup, name is : LAST and 1 to 5.
The LAST register holds the setup at power down.
First, you must verify if feature PWR ON AUTO is activated (ON). Under STORE/RECALL menu.
Second, you must verify that the name of register used at power up is LAST and not other name like 1 to 5. (On STORE/RECALL menu, under PWR ON AUTO)
I don't have a 34450A, but I suppose that feature is similar and avalaible for the 34450A.
Regards