I currently am using a N9030A PXA and am getting varying differences in readings when I use Average Detector and EMI Average Detector. Can someone please explain the difference of these two setting to me.
Thanks
I currently am using a N9030A PXA and am getting varying differences in readings when I use Average Detector and EMI Average Detector. Can someone please explain the difference of these two setting to me.
Thanks
From N9060-90027:
The Average detector determines the average of the signal within the bucket. The averaging method depends upon Average Type selection (voltage, power or log scales).
The EMI-Average detector provides a standard means to “smooth” the signal while still providing compliance to CISPR pulse response standards. It displays the average value of the amplitude envelope, rather than the average value of sample-detected amplitude, and uses an advanced algorithm to realize a lowpass filter that conforms to the latest CISPR standard.
From N9060-90027:
The Average detector determines the average of the signal within the bucket. The averaging method depends upon Average Type selection (voltage, power or log scales).
The EMI-Average detector provides a standard means to “smooth” the signal while still providing compliance to CISPR pulse response standards. It displays the average value of the amplitude envelope, rather than the average value of sample-detected amplitude, and uses an advanced algorithm to realize a lowpass filter that conforms to the latest CISPR standard.