I am recently building a VNA-FMR using the Agilent PNA-X N5245A tool. I'd like the analyzer does the frequency swep between certain values, and measure S21 at each frequency. To sweep the frequency, I found out an example file from Keysight website as below, and I modified based on it with LabVIEW 2017.
However, when I run the program, there is always an error message on the screen, as shown below. It says: The SCPI command received has caused error number -108 "Parameter not allowed".
I checked the program in labview when it runs, it seems that the error happens at a certain sub vi, which is shown in the next picture (Configure frequency start stop.vi):
This sub vi description is shown here as well:
I really couldn't get the error point and don't know how to solve it. When I use the example program shown in the beginning without any change, the same error happens at the same position in the program.
I am new to LabVIEW and its control to the analyzer. It might be a sily question, but I still look forward to your kindly help, since it confused me for several days. I also attached my program for your reference.
Thanks a lot for your help!
PS: I also found "labviewdriver_pna_8_2_11_2009" from the NI website. Should I use some subvis from this file or not?
Best regards
There should be a command error. You can use menu System>Configure>SICL>SCPI Monitor. Turn on the monitor to run the program, you'll see what commands have been sent to the PNA and which command caused the error. You can also save it to a text file.