What is the sampling speed for 34970A with 34901A if I scan one channel constantly to monitor one channel voltage?
This is in the data sheet.
Function: Thermocouple: Resolution 0.1C with 10 plc aperture (highest accuracy = slowest speed = greatest noise rejection): 6 rdgs/sec for 60 Hz power, 5 for 50 Hz. Resolution 0.1 C with 1 plc aperture: 52/47 rdgs/sec, Resolution 1 C, 0.02 plc: 280 rdgs/sec.
For voltage measurement 34901A , what is the sampling speed for 1 channel only in scan list?
This is in the data sheet.