Hi everyone,
I am Burak Telbakiroğlu. I am a student at the university. We have a power supply which mark is Agilent Technologies and its model is N6700B. I am trying a coating with electrodeposition by using this power supply. I want to change some of the parameters such as current applying and stopping. For example; assuming current will constant and its value is 10A but appliying time for 10 ms and after it stops for 5 ms. Again it gives 10A for 10 ms after it stops 5 ms for (4 hours: total time with this condition). So as to do this I installed Agilent Connection Expert and after opened this I also opened Keysight Command Expert program. However, I did not known the required codes to work power supply by this condition. Can anyone help to me? When I open the Keysight Command Expert, I had this page but I do not know necessary codes.
Best Regards.
Hi burkiburak,
You may refer to the SCPI commands in the Keysight Series N6700 Low-Profile Modular Power System Programmer’s Reference Guide from Page 206 to 210. Take note that the list mode is only available with models N676xA, N678xA SMU, or models that have Option 054 (High Speed Test Extensions).
The link to the reference guide is provided as follows:
Keysight Series N6700 Low-Profile Modular Power System Programmer's Reference Guide
For your information, there is an example on how to program an output pulse or pulse train in the Keysight Series N6700 Low-Profile Modular Power System User’s Guide on Page 73. Take note that the example is a voltage pulse. For your case, just change it the current.
The link to the user's guide is provided as follows:
Keysight Series N6700 Low-Profile Modular Power System User's Guide