On MSOX3014T when saving acquisition in H5 format ( or getting data using 'Waveform:Data?' ) increment between points is different from Analog signal and digital signal ?
Is it normal ? both analog and digital use same timebase
Also number of points are different
In my case :
analog signals : 16000 points with increment Xinc 3.12e-5
digital signal : 16666 points with increment Xinc 3.00012e-5
Was suprised to have different increment between analog and digital voices using same timebase.
The analog and digital sections sample separately. They are connected to the same timebase, and are synchronized at the trigger point, but they capture at their own rates, and the depth of the digital section is set to capture the same amount of time as the analog section.
Al