I have tried to perform simple 1-port waveguide calibration with E5071C analyzer. When I have measured short , offset short and fixed load "done" button stays still grey and it can't select. Why so, even I have measured all needed components? I have never had such an issue with coaxial calibrations. Should I do something before waveguide calibration on the analyzer settings or what? In the end I tried to test X/R/K kit functions and same problem with them too.
Hello,
Wave guide cal kit has its minimum and max frequency range that can be used, and your symptom seems you set wider frequency range on the ENA than that is covered by your cal kit. Please check your start & stop frequency, and frequency range that your wave guide cal kit covers. You cannot press done until you finish measruing all frequency range.
Cal kit freq. range can be confirmed by modify cal > define standard > select standard menu.