Unit: MSO-X 4154A
O-scope is not connected to DUT. BNCs are floating (no shorts on hand). Signal generator option was 50 ohm terminated and unterminated when problem was seen.
Background:
A few weeks ago (early June), went to a different site to collaborate with a new group on some hardware development. I took my o-scope with me as I have been using it as my means of recording RF waveforms in case I needed to perform sanity checking. I kept seeing extreme noise triggering my scope (example cell phone photo):
I then returned to my office and no longer saw the issue. In fact, for months prior, I did not see the issue.
My employer moved to a new office and I have been re-establishing my lab at the new location. Today, another in my lab saw this behavior with the o-scope. I took several photos, disconnected our DUT, switched inputs back to 1M probes (not connected) instead of the 50 ohm BNC cables we typically work with, and saw the problem multiple times:
So, is this earth ground with extreme bouncing? Is this something going on in the scope? How to I solve/mitigate this? This is causing us to trigger at the wrong time because we're triggering on this issue instead of the RF we're trying to capture.
It looks like a scope problem, especially since it appears to be happening on all channels at once.
Does it happen all the time?
What if you jiggle the scope a little, does that make it worse?
Does it happen on all V/div settings?
Try it with some 50 Ohm shorts.
You probably need to call your local Technical Call Center to see about getting it repaired.
Al