Hi,
I am using Agilent N6705B DC Power Analyser for current measurement. I want to transfer the logged data from Internal memory to External memory.
Problem:
While copying file from Internal to External memory, error message received: “Insufficient space in the destination drive ”.
Same error observed while exporting the logged data to External.
In short External memory is not detected.
Can you please provide solution to solve this issue?
Many thanks in advance!
Regards,
Prabha Chauhan
Hi prabhachauhan,
There are a few limitations for exporting data from the N6705B to external memory. You may want to take note of the following:
1. Exporting data from N6705B to any external memory requires a conversion from a .dlog file to a .csv file. The speed of the file conversion will be the limiting factor.
2. The external drive should be in FAT32 format. NTFS will not work.
If all else fails, here are some suggestions for you:
1. Try to use other external memory (e.g. with different memory size/ brand)
2. You can use the LXI webpage to transfer the data file from your unit to a PC