Dear Sir/Madam,
We purchased 4395A Network/Spectrum/Impedance analyzer and trying to measure the impedance of an Ultrasonic Transducer with 1meter long coaxial-cable. I kindly request you to please provide a calibration and measurement procedure to analyze the impedance of an ultrasonic transducer.
Thanking you.
With best regards,
Sanjeev
Sorry I don't know, but I do know there are a few pieces of information you would need to provide to get a useful answer.
1) What test set and other accessories do you have? The analyzer itself is not much use without appropriate accessories.
2) Do you have the option 010 - impedance analysis?
3) What cal kit(s).
4) What sort of connections do you have to your transducer?
5) Is the 1 m of cable attached to the transducer permanently?
There's a fair bit of useful information about the 4395A here
Agilent 4395A
The author of that site is quite knowledgeable on this, and might be able to give you some advice.
Unfortunately, there are no Keysight forums for LCR meter or impedance analyzers. Those instruments are built in Japan, and the staff in Japan rarely if ever read the forums. Questions about LCR meters and impedance analyzers tend to go unanswered.So I would not hold your breath waiting for an answer from Keysight staff on this.
Dave