Currently I would only like to ask you for reviewing the sequence of commands I am sending to the Agilent E5071C ENA. for a full 2 port calibration. I am trying to calibrate one only one channel. The exact sequence I am sending is the following. Is this sequence right?
I am considering moving the commands with stamps [1],[2] and [3] to the place where it says "HERE" as I understand that these parameters are set for a channel instead of for each trace. Am I right? I also would like to know if the command with the stamp [4] is necessary. I really do not know why it is there.
:DISP:SPL D1
:DISP:WIND1:ACT
HERE
:CALC1:PAR:COUN 4
:CALC1:PAR1:DEF S11
:CALC1:PAR1:SEL
:CALC1:SEL:FORM SCOM
:SENS1:FREQ:STAR 100e6                                [1]
:SENS1:FREQ:STOP 200e6                                [2]
:SENS1:SWE:POIN 30                                          [3]
:CALC1:PAR2:DEF S12
:CALC1:PAR2:SEL
:CALC1:PAR3:DEF S21
:CALC1:PAR3:SEL
:CALC1:PAR4:DEF S22
:CALC1:PAR4:SEL
:CALC1:PAR1:SEL                                               [4]
:SENS1:CORR:COLL:CKIT 30
:SENS1:CORR:COLL:CKIT:LABEL CS5
:SENS1:CORR:RVEL:COAX 0.442
:SENS1:CORR:COLL:CKIT:STAN1:LABEL CS5_OPEN
:SENS1:CORR:COLL:CKIT:STAN1:TYPE OPEN
:SENS1:CORR:COLL:CKIT:STAN1:C0 6.5e-15
:SENS1:CORR:COLL:CKIT:ORD:OPEN 1,1
:SENS1:CORR:COLL:CKIT:ORD:OPEN 2,1
:SENS1:CORR:COLL:CKIT:STAN2:LABEL CS5_SHORT
:SENS1:CORR:COLL:CKIT:STAN2:TYPE SHORT
:SENS1:CORR:COLL:CKIT:STAN2:L0 5e-12
:SENS1:CORR:COLL:CKIT:ORD:SHORT 1,2
:SENS1:CORR:COLL:CKIT:ORD:SHORT 2,2
:SENS1:CORR:COLL:CKIT:STAN3:LABEL CS5_LOAD
:SENS1:CORR:COLL:CKIT:STAN3:TYPE LOAD
:SENS1:CORR:COLL:CKIT:STAN3:DELAY 0.0156e-12
:SENS1:CORR:COLL:CKIT:STAN3:Z0 5
:SENS1:CORR:COLL:CKIT:ORD:LOAD 1,3
:SENS1:CORR:COLL:CKIT:ORD:LOAD 2,3
:SENS1:CORR:COLL:CKIT:STAN4:LABEL CS5_THRU
:SENS1:CORR:COLL:CKIT:STAN4:TYPE THRU
:SENS1:CORR:COLL:CKIT:STAN4:DELAY 1.13e-12
:SENS1:CORR:COLL:CKIT:ORD:THRU 1,2,4
:SENS1:CORR:COLL:METH:SOLT2 1, 2
*OPC?
:SENS1:CORR:COLL:OPEN 1
*OPC?
:SENS1:CORR:COLL:OPEN 2
*OPC?
*OPC?
:SENS1:CORR:COLL:SHORT 1
*OPC?
:SENS1:CORR:COLL:SHORT 2
*OPC?
*OPC?
:SENS1:CORR:COLL:LOAD 1
*OPC?
:SENS1:CORR:COLL:LOAD 2
*OPC?
*OPC?
:SENS1:CORR:COLL:THRU 1,2
*OPC?
:SENS1:CORR:COLL:THRU 2,1
*OPC?
*OPC?
:SENS1:CORR:COLL:SAVE
Thank you very much for your help