Hai all,
I have specific task of using the edge then edge trigger and store the wave forms for offline analysis.
Have two time-correlated random pulse trains into ch 1 & 2. I arm with ch1 and trigger with edge then edge trigger mode on event 0. When I use quick save using multipurpose trigger task, I was expecting each set/file to have pulse in both channels. How ever, most of the sets/files have only channel 1 pulse while channel 2 is baseline (thresholds are well above noise).
Is there any thing I miss in using this trigger mode to capture & save only when both chennels have one pulse each?
Thanks in advance
-rajaraman
rraman64,
I think your solution is in Sequence trigger. The title of this discussion says you have a 90000A. This should have Sequence trigger.
The setup below will trigger on ch1 falling edge, then ch2 falling edge. If no trigger occurs in 100ns, it resets and starts looking for ch1 trigger again. This should trigger on Case 2 only.
Sorry I did not pick up on this sooner.