How do I set/program the "System Service Calibration Date" and "Unit Serial #"?
I am trying to calibrate my HP 8546 and HP85460.
The "Last Cal" menu shows the date at 2000 and serial numbers at 225 and 110.
I am able to calibrate the system and use, but after a power cycle the calibration must be repeated.
Regards,
Dan
I am trying to calibrate my HP 8546 and HP85460.
The "Last Cal" menu shows the date at 2000 and serial numbers at 225 and 110.
I am able to calibrate the system and use, but after a power cycle the calibration must be repeated.
Regards,
Dan
Do you think that the 225 and 110 serial numbers are incorrect? Do the serial numbers on the rear panels end in these numbers? From what I remember the serial numbers displayed on-screen do not include the serial prefix, just the sequencial part of the serial number on the rear panel. So this could be correct.
If the instrument front panel alignment will not survive a power cycle you probably need to change the CPU battery.
Best regards,
Mike