:MEASure:VAMPlitude accuracy for 3000 series

Question asked by jlanderson on Jan 26, 2015
Latest reply on Jan 27, 2015 by bsnow
vamp = vtop-vbase

Single measurement accuracy is:
DC vertical accuracy ±(DC vertical gain accuracy + DC vertical offset accuracy + 0.25% full scale)
DC vertical gain accuracy* ± 2.0% full scale**
DC vertical offset accuracy ± 0.1 div ± 2 mV ± 1% of offset setting

– Single cursor accuracy:
  ±(DC vertical gain accuracy + DC vertical offset accuracy + 0.25% full scale) (same as single measurement)
– Dual cursor accuracy:
  ±(DC vertical gain accuracy + 0.5% full scale)*

I do not think cursor accuracy can be applied to a digitized waveform that does not use cursors to extract vtop and vbase.
The waveform is a single pulse referenced to zero.  No offset is used.
What is the correct accuracy equation to use?