Dear All,
I have bought new 53210A, and tried to write and read with Agilent Interactive IO, the result is:
* Connected to: USB0::0x0957::0x1707::MY50002282::0::INSTR
-> *IDN?
<- Agilent Technologies,53210A,MY50002282,01.12-1313.3321-1.19-4.10-127-154-35
But, I saw at 53210a display : "ERR", what is the means of "ERR" and how to fix this ?
I have tried example at folder "Basic_Measurement" with vc# 2008 express, when I press connect button, the result is "Agilent Technologies,53210A,MY50002282,01.12-1313.3321-1.19-4.10-127-154-35", I think the communication is good, but I have problem when press Make Measurement:
1. "ERR" at 53210A
2. VI_ERROR_TMO: A timeout occurred
How to fix this error? please share
I have bought new 53210A, and tried to write and read with Agilent Interactive IO, the result is:
* Connected to: USB0::0x0957::0x1707::MY50002282::0::INSTR
-> *IDN?
<- Agilent Technologies,53210A,MY50002282,01.12-1313.3321-1.19-4.10-127-154-35
But, I saw at 53210a display : "ERR", what is the means of "ERR" and how to fix this ?
I have tried example at folder "Basic_Measurement" with vc# 2008 express, when I press connect button, the result is "Agilent Technologies,53210A,MY50002282,01.12-1313.3321-1.19-4.10-127-154-35", I think the communication is good, but I have problem when press Make Measurement:
1. "ERR" at 53210A
2. VI_ERROR_TMO: A timeout occurred
How to fix this error? please share
The time out occurred because the error caused the counter to not complete the measurement within the timeout period set for Interactive IO. (This setting is located under the Interactive IO "Options" key.)
The error message should describe what caused the error. Once the condition causing the error is corrected, the time out should no longer occur.