The scope incorrectly sets 10:1 for the probe's scaling. Running "Probe Check" returns a message "...detected the probe to be 5:0 press OK to apply this value". Pressing OK instead of selecting 5:0 (which is correct for this probe) seems to run the probe check again and returns a message "...detected the probe to be 2:0 press OK to apply this value". Pressing OK again runs it yet again and returns a message "...detected the probe to be 1:0 press OK to apply this value". Pressing OK again briefly flashes a message which I can't read fast enough. The scaling remains on auto 10:1.
The same behaviour has been observed with the system firmware 2.35 and 2.12.
The only workaround is to insulate the probe's autodetect pin and manually setting the scaling to 5:0. Running the probe check then returns a pass.
Thanks
tomasz.
The same behaviour has been observed with the system firmware 2.35 and 2.12.
The only workaround is to insulate the probe's autodetect pin and manually setting the scaling to 5:0. Running the probe check then returns a pass.
Thanks
tomasz.