Hello,my instrument is blocked when i make a low voltage,so i think to the trigger edge but he probleme is not solved by the trigger edge and the scope is blocked and the trigger icon will flashes in the screen of the oscilloscope
i was chhanged the value of trigger level for many value(0,0.00001,1,.........) but the problem is the same ,can some help me :)
Edited by: omartun on Mar 6, 2014 2:00 PM
i was chhanged the value of trigger level for many value(0,0.00001,1,.........) but the problem is the same ,can some help me :)
Edited by: omartun on Mar 6, 2014 2:00 PM
The signal MUST go through some minimum screen height to trigger:
Trigger sensitivity: < 10 mV/div: greater of 1 div or 5 mV; ≥ 10 mV/div: 0.6 div
The ONLY way to trigger on DC is ion AUTO trigger mode, not NORMAL.
You can also use the :TRIGger:FORCe command if it's DC.