hi friends:
i have a N5182A unit.when i use NI GPIB explorer,i can find its address,but when it connects with my SA.
i will get a error address like the attachements.and i can not run ATE now
PS:my N5182A always has a error message after bootup,"700 state save recall Error,fail to recall the instrument state"
i have a N5182A unit.when i use NI GPIB explorer,i can find its address,but when it connects with my SA.
i will get a error address like the attachements.and i can not run ATE now
PS:my N5182A always has a error message after bootup,"700 state save recall Error,fail to recall the instrument state"
I assume you are trying to connect the MXG-A to a spectrum analyzer to perform an I/Q Mod Skew Cal.
In the users guide on page 99 the I/Q Mod Skew Cal has the procedure for connecting to the spectrum analyzer.
http://cp.literature.agilent.com/litweb/pdf/N5180-90003.pdf
As for the 700 error, you should set your power up to the preset condition. This can be selected under the Utility, Power on Preset, Power On, Preset.
Hope this was helpful -