Hello everybody,
I have a problem to want to ask everybody.
I have a agilent 53132a in the test system, It's error about the GPIB port, The computer can't control and communication with it,
I scare the GPIB port broken.
Can you help me, how can I repair or check it to know it broken or not.
Thank you very much.
I have a problem to want to ask everybody.
I have a agilent 53132a in the test system, It's error about the GPIB port, The computer can't control and communication with it,
I scare the GPIB port broken.
Can you help me, how can I repair or check it to know it broken or not.
Thank you very much.
What is the error number the instrument is reporting?
Have you cycled power on the instrument and checked that the GPIB cable is firmly connected to the GPIB connector on the instrument?
As for communicating with the counter from the PC, are you using the Agilent IO Libraries?
http://www.home.agilent.com/agilent/software.jspx?cc=US&lc=eng&nid=-33330.977662&id=2175637
Using the libraries' Agilent Connection Expert utility, is the utility able to locate and identify the instrument? If not, there may be a failure with the GPIB connector or with the circuitry supporting the GPIB interface. If an IO connection cannot be made, the instrument should be returned to an Agilent Service Center for repair.