I'd like to make here a question about an error that appear on my DSO6012A time to time just when turned on in the morning.
It gives me the message:
System concerns detected:
- The default setup was loaded
- Saved probe factors were defaulted
I updated oscilloscope to last software version and opened it, I measured battery inside and it seems ok.
What can be happened ? Error appears randomly only at power on. Is there something I can check to resolve this problem ?
I attach photo of this error. Thanks in advance for any help.
Greetings, Devis
I'm new on this forum.
I'd like to make here a question about an error that appear on my DSO6012A time to time just when turned on in the morning.
It gives me the message:
System concerns detected:
- The default setup was loaded
- Saved probe factors were defaulted
I updated oscilloscope to last software version and opened it, I measured battery inside and it seems ok.
What can be happened ? Error appears randomly only at power on. Is there something I can check to resolve this problem ?
I attach photo of this error. Thanks in advance for any help.
Greetings, Devis
http://www.home.agilent.com/agilent/editorial.jspx?cc=US&lc=eng&ckey=666864&nid=-536902766.0.00&id=666864
Make sure you grab the files for the 5000 and 6000, not the 7000 series…
do in order of:
sys
lib
gsp
lpk
If it still comes back, run a self test... if the slef test fails, run it again. If the self est keeps failing -> repair at Agilent service center. You could also do a self cal... if i passes, the scope acquisition hardware is (should be) good... then you just have to worry about starting from defualt every time... you can save setups internally or to a USB stick...