Update..........
When i try to increase the voltage the overrange led ligth and i've no uotput. The voltage increase circa 100mV and then stop with overrange status.
I've found a primari current sensor T!,i think that is a hall effect transducer that monitors the cerrent in the primati of the power transformer.
The signal of this transducer is linked to the power limit circuit in the control board.
Now if this transducer is ko probabli could give a signal of max current(the transducer should give a tension with the current flow)and give the supply a overrange status.
I can't find dat on
Primary-Current Monitor
Transformer T1 code
9100-4350 .
What do yuo think and do you know the spec of this transducer?
Thanks
Max
When i try to increase the voltage the overrange led ligth and i've no uotput. The voltage increase circa 100mV and then stop with overrange status.
I've found a primari current sensor T!,i think that is a hall effect transducer that monitors the cerrent in the primati of the power transformer.
The signal of this transducer is linked to the power limit circuit in the control board.
Now if this transducer is ko probabli could give a signal of max current(the transducer should give a tension with the current flow)and give the supply a overrange status.
I can't find dat on
Primary-Current Monitor
Transformer T1 code
9100-4350 .
What do yuo think and do you know the spec of this transducer?
Thanks
Max
Hello
Sorry that I am asking so many questions:
When was the instrument last working correctly?
When was the instrument last calibrated?
What is the acMains input voltage?
Are internal voltage selection switches set to correct acMains?
Is internal fuse good?
Is internal jumper set to correct Mains?
Does 6030A have Option 100?
What is the s/n of the instrument?
Are ALL internal jumper cables registered correctly in respective socket?
Are mode switches B1 – B6 ALL set correctly?
Are +/- sensing wires correctly in place?
To what value is OVP set?