Controling the Scope through Visual Basic
Question asked by hezi on Dec 24, 2012
Hi
I"m using dsox3034a scope and try to write script at VB Code.
I"m using "Programming Examples" from your Programmer's guide.
From some reason when I give "signal" command, the scope move to "signal" state but after ~2-3 seconds it back to "noraml" state and not waiting for the signal as I exept and as it act when I use manual.
I will be very happy if you tell me were I wrong.
Thanks
Hezi
Attached VB code.
the code:
' Agilent VISA Example in Visual Basic .NET
' This program illustrates a few commonly-used programming
' features of your Agilent oscilloscope.
Imports System
Imports System.IO
Imports System.Text
Namespace InfiniiVision
    Class VisaInstrumentApp
        Private Shared myScope As VisaInstrument
        Public Shared Sub Main(ByVal args As String())
            Try
                myScope = New VisaInstrument("USB0::2391::6052::my52012979::0::INSTR")
                myScope.SetTimeoutSeconds(10)
                ' Initialize - start from a known state.
                Initialize()
                ' Capture data.
                Capture()
                ' Analyze the captured waveform.
                Analyze()
            Catch err As System.ApplicationException
                Console.WriteLine("*** VISA Error Message : " + err.Message)
            Catch err As System.SystemException
                Console.WriteLine("*** System Error Message : " + err.Message)
            Catch err As System.Exception
                Debug.Fail("Unexpected Error")
                Console.WriteLine("*** Unexpected Error : " + err.Message)
            End Try
            Console.ReadLine()
        End Sub
        ' Initialize the oscilloscope to a known state.
        Private Shared Sub Initialize()
            Dim strResults As StringBuilder
            ' Get and display the device's *IDN? string.
            strResults = myScope.DoQueryString("*IDN?")
            Console.WriteLine("*IDN? result is: {0}", strResults)
            'Clear status and load the default setup.
            myScope.DoCommand("*CLS")
            myScope.DoCommand("*RST")
        End Sub
        ' Capture the waveform.
        Private Shared Sub Capture()
            ' Use auto-scale to automatically configure oscilloscope.
            'myScope.DoCommand(":AUToscale")
            myScope.DoCommand(":SINGle")
            ' Set trigger mode (EDGE, PULSe, PATTern, etc., and input source.
            myScope.DoCommand(":TRIGger:MODE EDGE")
            Console.WriteLine("Trigger mode: {0}", myScope.DoQueryString(":TRIGger:MODE?"))
            ' Set EDGE trigger parameters.
            myScope.DoCommand(":TRIGger:EDGE:SOURCe CHANnel1")
            Console.WriteLine("Trigger edge source: {0}", myScope.DoQueryString(":TRIGger:EDGE:SOURce?"))
            myScope.DoCommand(":TRIGger:EDGE:LEVel 1")
            Console.WriteLine("Trigger edge level: {0}", myScope.DoQueryString(":TRIGger:EDGE:LEVel?"))
            myScope.DoCommand(":TRIGger:EDGE:SLOPe NEGative")
            Console.WriteLine("Trigger edge slope: {0}", myScope.DoQueryString(":TRIGger:EDGE:SLOPe?"))
            ' Save oscilloscope configuration.
            Dim ResultsArray As Byte() ' Results array.
            Dim nLength As Integer ' Number of bytes returned from inst.
            Dim strPath As String
            Dim fStream As FileStream
            ' Query and read setup string.
            nLength = myScope.DoQueryIEEEBlock(":SYSTem:SETup?", ResultsArray)
            ' Write setup string to file.
            'strPath = "C:\scope\config\setup.stp"
            'fStream = File.Open(strPath, FileMode.Create)
            'fStream.Write(ResultsArray, 0, nLength)
            'fStream.Close()
            'Console.WriteLine("Setup bytes saved: {0}", nLength)
            ' 'Change settings with individual commands:
            '' Set vertical scale and offset.
            myScope.DoCommand(":CHANnel1:SCALe 3 ")
            Console.WriteLine("Channel 1 vertical scale: {0}", myScope.DoQueryString(":CHANnel1:SCALe?"))
            ' myScope.DoCommand(":CHANnel1:OFFSet -1.5")
            ' Console.WriteLine("Channel 1 vertical offset: {0}", myScope.DoQueryString(":CHANnel1:OFFSet?"))
            '' Set horizontal scale and position.
            myScope.DoCommand(":TIMebase:SCALe 0.5")
            Console.WriteLine("Timebase scale: {0}", myScope.DoQueryString(":TIMebase:SCALe?"))
            myScope.DoCommand(":TIMebase:POSition 0.0")
            Console.WriteLine("Timebase position: {0}", myScope.DoQueryString(":TIMebase:POSition?"))
            ' Set the acquisition type (NORMal, PEAK, AVERage, or HRESolution)
            myScope.DoCommand(":ACQuire:TYPE NORMal")
            Console.WriteLine("Acquire type: {0}", myScope.DoQueryString(":ACQuire:TYPE?"))
            ' Or, configure by loading a previously saved setup.
            Dim DataArray As Byte()
            Dim nBytesWritten As Integer
            ' Read setup string from file.
            'strPath = "c:\scope\config\setup.stp"
            'DataArray = File.ReadAllBytes(strPath)
            ' Restore setup string.
            'nBytesWritten = myScope.DoCommandIEEEBlock(":SYSTem:SETup", DataArray)
            'Console.WriteLine("Setup bytes restored: {0}", nBytesWritten)
            ' Capture an acquisition using :DIGitize.
            'myScope.DoCommand(":DIGitize CHANnel1")
        End Sub
        Private Shared Sub Analyze()
            Dim fResult As Double
            Dim pResult As Double
            Dim nwResult As Double
            Dim pwResult As Double
            Dim Coun As Integer
            ' Make a couple of measurements.
            'myScope.DoCommand(":MEASure:SOURce CHANnel1")
            'Console.WriteLine("Measure source: {0}", myScope.DoQueryString(":MEASure:SOURce?"))
            'myScope.DoCommand(":MEASure:FREQuency")
            'fResult = myScope.DoQueryNumber(":MEASure:FREQuency?")
            'Console.WriteLine("Frequency: {0:F4} kHz", fResult / 1000)
            'myScope.DoCommand(":MEASure:VAMPlitude")
            'fResult = myScope.DoQueryNumber(":MEASure:VAMPlitude?")
            'Console.WriteLine("Vertial amplitude: {0:F2} V", fResult)
            'myScope.DoCommand(":MEASure:PERiod")
            'pResult = myScope.DoQueryNumber(":MEASure:PERiod?")
            'Console.WriteLine("Period amplitude: {0:F2} V", pResult)
            ' myScope.DoCommand(":MEASure:NWIDth")
            'nwResult = myScope.DoQueryNumber(":MEASure:NWIDth?")
            ' Console.WriteLine("Nwidth amplitude: {0:F2} V", nwResult)
            myScope.DoCommand(":MEASure:PWIDth")
            pwResult = myScope.DoQueryNumber(":MEASure:PWIDth?")
            Console.WriteLine("Nwidth amplitude: {0:F2} V", pwResult)
            Coun = 40000000
            Console.WriteLine("Coun = {0:D}", Coun)
            While Coun > 0
                Coun = Coun - 1
                If (Coun = 1) Then Console.WriteLine("Coun = 1")
                'Console.WriteLine("Coun = ", Coun)
            End While
            ' Download the screen image.
            myScope.DoCommand(":HARDcopy:INKSaver OFF")
            ' Get the screen data.
            nLength = myScope.DoQueryIEEEBlock(":DISPlay:DATA? PNG, COLor", ResultsArray)
            ' Store the screen data to a file.
            strPath = "c:\scope\data\screen.png"
            fStream.Close()
            Console.WriteLine("Screen image ({0} bytes) written to {1}", nLength, strPath)
            ' Download waveform data.
            ' ' Set the waveform points mode.
            '   myScope.DoCommand(":WAVeform:POINts:MODE RAW")
            '  Console.WriteLine("Waveform points mode: {0}", myScope.DoQueryString(":WAVeform:POINts:MODE?"))
            '' Get the number of waveform points available.
            'myScope.DoCommand(":WAVeform:POINts 10240")
            'Console.WriteLine("Waveform points available: {0}", myScope.DoQueryString(":WAVeform:POINts?"))
            '' Set the waveform source.
            'myScope.DoCommand(":WAVeform:SOURce CHANnel1")
            'Console.WriteLine("Waveform source: {0}", myScope.DoQueryString(":WAVeform:SOURce?"))
            '' Choose the format of the data returned (WORD, BYTE, ASCII):
            'myScope.DoCommand(":WAVeform:FORMat BYTE")
            'Console.WriteLine("Waveform format: {0}", myScope.DoQueryString(":WAVeform:FORMat?"))
            ' Display the waveform settings:
            Dim fResultsArray As Double()
            fResultsArray = myScope.DoQueryNumbers(":WAVeform:PREamble?")
            Dim fFormat As Double = fResultsArray(0)
            If fFormat = 0 Then
                Console.WriteLine("Waveform format: BYTE")
            ElseIf fFormat = 1 Then
                Console.WriteLine("Waveform format: WORD")
            ElseIf fFormat = 2 Then
                Console.WriteLine("Waveform format: ASCii")
            End If
            Dim fType As Double = fResultsArray(1)
            If fType = 0 Then
                Console.WriteLine("Acquire type: NORMal")
            ElseIf fType = 1 Then
                Console.WriteLine("Acquire type: PEAK")
            ElseIf fType = 2 Then
                Console.WriteLine("Acquire type: AVERage")
            ElseIf fType = 3 Then
                Console.WriteLine("Acquire type: HRESolution")
            End If
            Dim fPoints As Double = fResultsArray(2)
            Console.WriteLine("Waveform points: {0:e}", fPoints)
            Dim fCount As Double = fResultsArray(3)
            Console.WriteLine("Waveform average count: {0:e}", fCount)
            Dim fXincrement As Double = fResultsArray(4)
            Console.WriteLine("Waveform X increment: {0:e}", fXincrement)
            Dim fXorigin As Double = fResultsArray(5)
            Console.WriteLine("Waveform X origin: {0:e}", fXorigin)
            Dim fXreference As Double = fResultsArray(6)
            Console.WriteLine("Waveform X reference: {0:e}", fXreference)
            Dim fYincrement As Double = fResultsArray(7)
            Console.WriteLine("Waveform Y increment: {0:e}", fYincrement)
            Dim fYorigin As Double = fResultsArray(8)
            Console.WriteLine("Waveform Y origin: {0:e}", fYorigin)
            Dim fYreference As Double = fResultsArray(9)
            Console.WriteLine("Waveform Y reference: {0:e}", fYreference)
            ' Get the waveform data.
            nLength = myScope.DoQueryIEEEBlock(":WAVeform:DATA?", ResultsArray)
            Console.WriteLine("Number of data values: {0}", nLength)
            ' Set up output file:
            strPath = "c:\scope\data\waveform_data.csv"
            If File.Exists(strPath) Then
                File.Delete(strPath)
            End If
            ' Open file for output.
            Dim writer As StreamWriter = File.CreateText(strPath)
            ' Output waveform data in CSV format.
            For index As Integer = 0 To nLength - 1
                ' Write time value, voltage value.
                writer.WriteLine("{0:f9}, {1:f6}", fXorigin + (CSng(index) * fXincrement), ((CSng(ResultsArray(index)) - fYreference) * fYincrement) + fYorigin)
            Next
            ' Close output file.
            writer.Close()
            Console.WriteLine("Waveform format BYTE data written to {0}", strPath)
        End Sub
    End Class
        Private m_nResourceManager As Integer
        Private m_nSession As Integer
        Private m_strVisaAddress As String
        ' Constructor.
        Public Sub New(ByVal strVisaAddress As String)
            ' Save VISA addres in member variable.
            m_strVisaAddress = strVisaAddress
            ' Open the default VISA resource manager.
            OpenResourceManager()
            ' Open a VISA resource session.
            OpenSession()
            ' Clear the interface.
            Dim nViStatus As Integer
            nViStatus = visa32.viClear(m_nSession)
        End Sub
        Public Sub DoCommand(ByVal strCommand As String)
            ' Send the command.
            VisaSendCommandOrQuery(strCommand)
            ' Check for inst errors.
            CheckInstrumentErrors(strCommand)
        End Sub
        Public Function DoCommandIEEEBlock(ByVal strCommand As String, ByVal DataArray As Byte()) As Integer
            ' Send the command to the device.
            Dim strCommandAndLength As String
            nLength = DataArray.Length
            strCommandAndLength = [String].Format("{0} #8{1:D8}", strCommand, nLength)
            ' Write first part of command to formatted I/O write buffer.
            nViStatus = visa32.viPrintf(m_nSession, strCommandAndLength)
            CheckVisaStatus(nViStatus)
            ' Write the data to the formatted I/O write buffer.
            nViStatus = visa32.viBufWrite(m_nSession, DataArray, nLength, _
            nBytesWritten)
            Return nBytesWritten
        End Function
        Public Function DoQueryString(ByVal strQuery As String) As StringBuilder
            ' Send the query.
            VisaSendCommandOrQuery(strQuery)
            ' Get the result string.
            Dim strResults As New StringBuilder(1000)
            strResults = VisaGetResultString()
            CheckInstrumentErrors(strQuery)
            ' Return string results.
            Return strResults
        End Function
        Public Function DoQueryNumber(ByVal strQuery As String) As Double
            ' Send the query.
            Dim fResults As Double
            fResults = VisaGetResultNumber()
            Return fResults
        End Function
        Public Function DoQueryNumbers(ByVal strQuery As String) As Double()
            ' Send the query.
            fResultsArray = VisaGetResultNumbers()
            ' Check for instrument errors (another command and result).
            Return fResultsArray
        End Function
        Public Function DoQueryIEEEBlock(ByVal strQuery As String, ByRef ResultsArray As Byte()) As Integer
            ' Send the query.
            System.Threading.Thread.Sleep(2000) ' Delay before reading data.
            Dim length As Integer
            ' Number of bytes returned from instrument.
            length = VisaGetResultIEEEBlock(ResultsArray)
            Return length
        End Function
        Private Sub VisaSendCommandOrQuery(ByVal strCommandOrQuery As String)
            ' Send command or query to the device.
            Dim strWithNewline As String
            strWithNewline = [String].Format("{0}" & Chr(10) & "", _
            strCommandOrQuery)
            nViStatus = visa32.viPrintf(m_nSession, strWithNewline)
        End Sub
        Private Function VisaGetResultString() As StringBuilder
            ' Read return value string from the device.
            nViStatus = visa32.viScanf(m_nSession, "%1000t", strResults)
            Return strResults
        End Function
        Private Function VisaGetResultNumber() As Double
            Dim fResults As Double = 0
            nViStatus = visa32.viScanf(m_nSession, "%lf", fResults)
            Return fResults
        End Function
        Private Function VisaGetResultNumbers() As Double()
            fResultsArray = New Double(9) {}
            nViStatus = visa32.viScanf(m_nSession, "%,10lf" & Chr(10) & "", fResultsArray)
        End Function
        Private Function VisaGetResultIEEEBlock(ByRef ResultsArray As Byte()) As Integer
            ' Results array, big enough to hold a PNG.
            ResultsArray = New Byte(299999) {}
            ' Set the default number of bytes that will be contained in
            ' the ResultsArray to 300,000 (300kB).
            length = 300000
            nViStatus = visa32.viScanf(m_nSession, "%#b", length, ResultsArray)
            ' Write and read buffers need to be flushed after IEEE block?
            nViStatus = visa32.viFlush(m_nSession, visa32.VI_WRITE_BUF)
            nViStatus = visa32.viFlush(m_nSession, visa32.VI_READ_BUF)
            Return length
        End Function
        Private Sub CheckInstrumentErrors(ByVal strCommand As String)
            ' Check for instrument errors.
            Dim strInstrumentError As New StringBuilder(1000)
            Dim bFirstError As Boolean = True
            Do ' While not "0,No error"
                VisaSendCommandOrQuery(":SYSTem:ERRor?")
                strInstrumentError = VisaGetResultString()
                If Not strInstrumentError.ToString().StartsWith("+0,") Then
                    If bFirstError Then
                        Console.WriteLine("ERROR(s) for command '{0}': ", strCommand)
                        bFirstError = False
                    End If
                    Console.Write(strInstrumentError)
                End If
            Loop While Not strInstrumentError.ToString().StartsWith("+0,")
        End Sub
        Private Sub OpenResourceManager()
            nViStatus = visa32.viOpenDefaultRM(Me.m_nResourceManager)
            If nViStatus < visa32.VI_SUCCESS Then
                Throw New  _
                ApplicationException("Failed to open Resource Manager")
            End If
        End Sub
        Private Sub OpenSession()
            nViStatus = visa32.viOpen(Me.m_nResourceManager, Me.m_strVisaAddress, visa32.VI_NO_LOCK, visa32.VI_TMO_IMMEDIATE, Me.m_nSession)
        End Sub
        Public Sub SetTimeoutSeconds(ByVal nSeconds As Integer)
            nViStatus = visa32.viSetAttribute(Me.m_nSession, visa32.VI_ATTR_TMO_VALUE, nSeconds * 1000)
        End Sub
        Public Sub CheckVisaStatus(ByVal nViStatus As Integer)
            ' If VISA error, throw exception.
                Dim strError As New StringBuilder(256)
                visa32.viStatusDesc(Me.m_nResourceManager, nViStatus, strError)
                Throw New ApplicationException(strError.ToString())
            End If
        End Sub
        Public Sub Close()
            If m_nSession <> 0 Then
                visa32.viClose(m_nSession)
            End If
            If m_nResourceManager <> 0 Then
                visa32.viClose(m_nResourceManager)
            End If
        End Sub
    End Class
End Namespace