Does Agilent provide an -H84 calibration for E9304A-H18 sensors? I ask because my manager wants to have that one sensor calibrated with the -H84 option instead of the two sensors (8482A and 8481A) we currently send, to save money of course. Although the E9304A-H18 covers the frequency range, and has better SWR, my argument against it is as follows:
1. The E9304A is a diode sensor, which are typically not as linear as thermocouple sensors.
2. It has a low-power path which will have higher uncertainties than the high-power path. These worse uncertainties will have to be used with any other power sensors calibrated by this -H84 standard sensor, if the low-power path is used.
Is my argument valid? Are there more reasons I should not get an -H84 option calibration for a diode sensor?
1. The E9304A is a diode sensor, which are typically not as linear as thermocouple sensors.
2. It has a low-power path which will have higher uncertainties than the high-power path. These worse uncertainties will have to be used with any other power sensors calibrated by this -H84 standard sensor, if the low-power path is used.
Is my argument valid? Are there more reasons I should not get an -H84 option calibration for a diode sensor?