It still depends on the instrument. There was a bug in the early HP/Agilent common parser where the interrupt for the input buffer was given a higher priority than the output buffer's interrupt. The net effect was that completing an enter on an instrument query, followed by sending the next command could generate a query interrupted error. This was because the instrument detected the first incoming byte before clearing the last interrupt out of the output buffer, ie the buffer empty interrupt. The error was corrected in the 1995 timeframe, and all subsequent instruments act correctly, but a good number of popular instruments such as the 34401 and 33120 are affected with this problem. We found it much safer to use the output buffer empty bit in the status byte and do a spoll to ascertain it was clear before sending the next command. You will find this sequence in the panel drivers and the VXIplug&play drivers from Agilent.
Further, there was never any compliance criteria or testing agency established for IEEE 488.2 no SCPI, thus there are a lot of implementations out there where the instrument designer simply didn't understand the nuances of I/O. This is especially true in the OPC implementations, as nobody seems to understand when an operation is complete, especially the overlapped operations. This is why we found the output buffer empty to be safest for any command following a query.
Jay Nemeth-Johannes
Smart Sensor Systems
720 SW 14th Street
Loveland, Colorado 80537
(970) 663-0006 (Office)
(970) 290-9797 (Cell)
www.SmartSensorSystems.com
> -----Original Message-----
> From: Baranski, Detlef [mailto:Detlef.Baranski@pilkington.de]
> Sent: Friday, December 07, 2007 3:57 AM
> To: VRF
> Subject: AW: Re[2]: [vrf] GPIB error --*OPC?
> Snippet from Agilent web site:
> Using *OPC?
> When you programmatically setup an instrument for an operation, you
> typically send multiple instrument commands.
> Most instruments are able to buffer multiple commands.
> Allowing the instrument to receive new commands, even though a
> previous command is not yet completed.
> This code sends several commands to an instrument.
> With DMM
> .WriteString "*RST"
> .WriteString "*CLS"
> .WriteString "*ESE 1"
> .WriteString "*SRE 32"
> End With
> Provided the instrument can complete all these commands before an
> additional conflicting command or query is sent, the code will work as
> anticipated. A better way to send a block of commands is to have the
> instrument report when the block is complete before sending additional
> commands. The SCPI standard provides the *OPC? query that you can use
> to ensure the instrument has completed all operations before your
> program continues.
> An instrument responds to the *OPC? query by setting a "1" in the
> output buffer after all received commands have completed.
> This allows you to synchronize your program and the instrument
> operation.
> With DMM
> .WriteString "*RST"
> .WriteString "*CLS"
> .WriteString "*OPC?" ' Assure syncronization
> strTemp = .ReadString ' Discard returned value
> End With
> Note that using the *OPC? query holds the I/O bus until the instrument
> commands have been completed. It may not be acceptable to hold the
> GPIB bus when you are controlling multiple instruments. In those cases
> use the *OPC bit in the Standard Event register and implement some
> form of status reporting system. An example using the *OPC bit is
> provided in the Using SRQ Events section.
> Related Links
> Synchronizing Instrument I/O with *OPC?, SRQ, and polling If you have
> worked through the other section in this series, you have probably
> come to realize that, in addition to exchanging data, you need control
> when the exchange takes place.
> -----Ursprüngliche Nachricht-----
> Von: Baranski, Detlef
> Gesendet: Freitag, 7. Dezember 2007 10:55
> An: 'Stan Bischof (Richard S)'; VRF
> Betreff: AW: Re[2]: [vrf] GPIB error
> Why not using the Operation Complete Command ( *OPC?) to make sure ?
> Von: Stan Bischof (Richard S) [mailto:rsb@soco.agilent.com]
> Gesendet: Donnerstag, 6. Dezember 2007 18:16
> An: VRF
> Cc: vrf@agilent.com
> Betreff: Re[2]: [vrf] GPIB error
> "Ed Shaw" <eshaw@comtechefdata.com> wrote:
> > Recently I've seen a RF power meter that will freeze(comm
> timeout if
> > you try to communicate with it) when the power to the
> sensor goes below a given level.
> > Then when the RF power going to the sensor is increased all is well
> > again. In this case my small delay between the query and
> read command
> > mentioned above was of no help.
> At low enough power levels some power meters can take a very long time
> to settle-- and this results in comm timeouts if you haven't set the
> timeout long enough.
> Check your power meter and see how long it takes to settle at the low
> power levels you refer to. It may well be longer than the timeout you
> have set in VEE.
> Stan
> Stan Bischof Agilent Technologies 707-577-3994
> stan_bischof@agilent.com
> You are currently subscribed to vrf as:
> Detlef.Baranski@pilkington.de To subscribe please send an email to:
> "vrf-request@lists.it.agilent.com" with the word subscribe in the
> message body.
> Search the "unofficial vrf archive" at
> "http://www.vrfarchive.com/vrf_archive".
> Search the Agilent vrf archive at
> "http://vee.engineering.agilent.com".
> Der Inhalt dieser E-Mail (inkl. aller Anlagen) ist vertraulich und
> ausschließlich für den Adressaten bestimmt.
> Jede Art der Verbreitung, Nutzung oder Vervielfältigung ist untersagt.
> Sollten Sie diese E-Mail irrtümlich erhalten haben, informieren Sie
> bitte sofort den Absender und löschen die E-Mail.
> Unternehmensangaben der Gesellschaften mit Sitz in Deutschland:
> Pilkington Holding GmbH, Haydnstraße 19, 45884 Gelsenkirchen
> Sitz: Gelsenkirchen, Amtsgericht Gelsenkirchen HRB 8997
> Geschäftsführung: Jochen Settelmayer (Vorsitz), Dr. Clemens Miller,
> Thomas Kretschmann, Robert Hales Vorsitzender des
> Aufsichtsrates: Dr. Axel Wiesener
> Pilkington Deutschland AG, Haydnstraße 19, 45884 Gelsenkirchen
> Sitz: Gelsenkirchen, Amtsgericht Gelsenkirchen HRB 2707
> Vorstand: Dr. Clemens Miller, Jochen Settelmayer Vorsitzender des
> Pilkington Automotive Deutschland GmbH, Otto-Seeling-Straße 7, 58455
> Witten
> Sitz: Witten, Amtsgericht Bochum HRB 8443
> Geschäftsführung: Thomas Kretschmann
> Vorsitzender des Aufsichtsrates: Jochen Settelmayer
> Bauglasindustrie GmbH, Hüttenstraße 33, 66839 Schmelz / Saar
> Sitz: Schmelz / Saar, Amtsgericht Saarbrücken HRB 52020
> Geschäftsführung: Manfred Ebbers
> jay.johannes@completetest.com To subscribe please send an email to:
> message body.
Jay Nemeth-Johannes
720 SW 14th Street
> To: VRF
> Using *OPC?
> With DMM
> .WriteString "*RST"
> .WriteString "*CLS"
> End With
> program continues.
> operation.
> With DMM
> .WriteString "*RST"
> .WriteString "*CLS"
> End With
> Related Links
> An: VRF
> Cc: vrf@agilent.com
> timeout if
> read command
> have set in VEE.
> Stan
> message body.
> Witten
> message body.