I would like to ask if is there any references or guide for troubleshooting our hp3070 Tester?
I encountered problem with our tester where in it fails in Setting up Power Supply for all the Fixture.Single or Double Bank
I already changed ASRU Card in Module 3 since it was failing in diagnostics(DUT Lead Resistance) but still come out this
Failure(2340-2341)
I Swapped DUT Supply at module 3 to the another Tester but still same.
And i noticed that when i test power supply without the Unit under test in the fixture ,it will still failed the same,and the value also same.
I hope you can help me with this matter thank you..
James
I would like to ask if is there any references or guide for troubleshooting our hp3070 Tester?
I encountered problem with our tester where in it fails in Setting up Power Supply for all the Fixture.Single or Double Bank
I already changed ASRU Card in Module 3 since it was failing in diagnostics(DUT Lead Resistance) but still come out this
Failure(2340-2341)
I Swapped DUT Supply at module 3 to the another Tester but still same.
And i noticed that when i test power supply without the Unit under test in the fixture ,it will still failed the same,and the value also same.
I hope you can help me with this matter thank you..
James
You can also try to test the power supply in the system diagnostics, and check that the fans are working ok on the DUT PS
or can try to communicate with them over GPIB/HPIB using the btio command in shell, this will ensure you can communicate to the power supply
also DUT lead resistance failure, just generally can be a dirty MINT pins, or poor connection to the ASRU card - especially if it just marginally fails the DGN