_*Crosstalk from fixture wires*_
Did you know that how you layout the wires in the fixture is very important?
In order to reduce cross talk between wires in a fixture, ideally you should run the wires perpendicular to each other.
Fixture wires running in a bunch, parallel to each other, may look neat and pleasing to the eye, but it is devastating for signal quality. This is because signals running on one wire is inductively coupled to the next parallel wire. The more surface area that the two wires are in parallel, the stronger the coupling, the stronger the coupling, the more likely the signal from one wire can affect the other, this manifests as noise on the other wire.
To minimize crosstalk interference, the wires in the fixture should not be bundled together and should be wired directly from one Pin to another in the shortest path so that the wires intersect each other instead of run in parallel. This reduces the surface area of the wires that are in parallel and thus reduces the effect of crosstalk.
It is actually easier to wire from pin to pin. This small step could help you reduce lots of time debugging a noisy intermittent test. Try it.
With BGAs being very commonly found on PCBA these days, it is important to know the best way to size your VTEP sensor plates that are used to test the BGAs to achieve the best results.
For ICs with leads (gull-wings, J-leads, bat-wings, etc), we generally size the VTEP sensor plate larger than the component body to overlap over the leads. This is so that the capacitance coupling covers a wider surface area (the wire-bond, internal IC traces, and leads) resulting in a larger capacitance detected. Consequently, it would be easier to determine a well defined PASS/FAIL result.
For BGAs, the layout of the balls is different from ICs with leads. The balls are found under the component body instead of at the perimeter of the component. Furthermore, the balls are much smaller in surface area than the leads. A much more sensitive measurement method like VTEP is required compared to testing leaded ICs. However, with this added sensitivity, additional care is required to remove potential “noise” sources in order to get the best test results.
For best results, VTEP sensor plate should be sized smaller than the BGA. This reduces the coupling of surrounding components and PCB traces and therefore reduces the parasitic capacitance from these interferences. The illustration shows a sensor plate that is too large for the BGA. You can see that stray capacitance is introduced from a nearby component (C Component) on the left and from the trace (C Trace) running at the right side of the BGA under test. The final measurement of this VTEP is the sum of the BGA measurement (C Measurement), the nearby component (C Component) and the trace (C Trace).
If the VTEP sensor plate is reduced in size to about 80% of the BGA size, it will be large enough to cover just the BGA balls (or even slightly smaller). This drastically reduces the stray capacitance from the nearby component (C Component) and from the trace (C Trace), leaving only what you desire to measure (C Measurement).
Try this simple adjustment to the sensor plates on your BGAs and enjoy a more consistent VTEP test for BGAs.