Hi,
I am currently use the Agilent E3631A Triple Output DC Power Supply and control its setting use RS232 remote interface.
Question - After I set the E3631A e.g. "APPL P25V,12.0,0.1" with constant current output, the power supply will remain in limit mode. How can I return to meter mode to display the actural volatge output reading? which commands I can be used?
Your helps is much appreciated.
Peggy
I am currently use the Agilent E3631A Triple Output DC Power Supply and control its setting use RS232 remote interface.
Question - After I set the E3631A e.g. "APPL P25V,12.0,0.1" with constant current output, the power supply will remain in limit mode. How can I return to meter mode to display the actural volatge output reading? which commands I can be used?
Your helps is much appreciated.
Peggy
Thanks for your advise. I do aware that "command SYSTem:LOCal" will switch the power supply back to local mode. However, In my application I need to lock the E3631a front panel during the testing cycle and not allow any change by the operators, but need to show them the actual voltage output under the Vcc mode so they can record down the voltage value.
Any advise in this subject is truly appreciated.
Peggy
It works. Thanks!
One last question. Now if Iwould like to read both voltage and current values from the E3631A comm port? which commands I should be used to trigger E3631A to export these data?
Thanks in advace.
Peggy