Russ McCrory mentioned some problems about reading error terms from the hp8720
network analyzer. I also had some strange effects with direct I/O objects doing
the same for hp8753* and hp8719/20 networkanalyzers. In the end I started changing
a driver that was available for the hp8753c in the id-language. All hp nwa's
act nearly the same over hpib. So I added an 'identity' component and based upon
that value I added some CASE statements in those rare places where hp8753a, -b,
-c, -d or hp8719c differ in command sytax. (In this way I made also a driver available
for the hp8719c since it is not supplied by Hewlett-Packard's VEE, that's why I
mention it.)
I have also added, among other features, more functionality to the 'Active_trace'
component in the driver to support the transfer of errorterms to/from the instrument.
That seems to work very well, at least I have not yet encoutered any problems compared
to direct I/O. I have encluded part of the driver in an 'un-documented' fashion. It
are the 'Data_Array' component and the 'Get_Data_Array' actions block. There you can
see how the '#A' blocktransfer is decoded. You could make a dedicated driver just
for the purpose of errorterm transfer based upon these components (?)
Regards
Remi Tuijtelaars
Philips Semiconductors, Dev. T&D, Building BY2.021,
Gerstweg 2, NL-6534 AE Nijmegen, Netherlands.
tel/fax: + 31 24 353 - 2812 / 3160.
email: Remi.Tuijtelaars@nym.sc.philips.com
(seri: nytuijte@nyhp02)
COMPONENT Data_Array NOTSAVED NOGEN NOERRCHECK;
TYPE RARRAY 2,1602;
INITIAL INVALID;
SET ACTIONS;
! SELECT Channel;
! CASE Channel1;
! GET Number_of_points1;
! CASE Channel2;
! GET Number_of_points2;
! END SELECT;
SELECT Active_Trace;
CASE (Active_Trace) Data;
OUTPUT STRING "HOLD;FORM3;INPUDATA;";
GOSUB Set_Data_Array;
GOSUB Set_sweep_mode;
CASE (Active_Trace) Form;
OUTPUT STRING "HOLD;FORM3;INPUFORM;";
CASE (Active_Trace) Raw_1;
GET Correction_Off_On;
FETCH (Correction_Off_On) Off;
FETCH Correction_Off_On; EQ;
IF STACK THEN; ! Need to turn correction ON to write raw data arrays
NOTIFY "Turn Correction ON To Write Raw Data Arrays.";
ELSE;
OUTPUT STRING "HOLD;FORM3;INPURAW1;";
END IF;
CASE (Active_Trace) Raw_2;
IF STACK THEN;
ELSE;
OUTPUT STRING "HOLD;FORM3;INPURAW2;";
END IF;
CASE (Active_Trace) Raw_3;
IF STACK THEN;
ELSE;
OUTPUT STRING "HOLD;FORM3;INPURAW3;";
END IF;
CASE (Active_Trace) Raw_4;
IF STACK THEN;
ELSE;
OUTPUT STRING "HOLD;FORM3;INPURAW4;";
END IF;
CASE (Active_Trace) Memory;
OUTPUT STRING "DISPMEMO;HOLD;FORM3;INPUDATA;";
OUTPUT STRING "WAIT;DATI;";
GOSUB Set_Display_Type; ! Reset the display back where it was ! check
CASE ELSE; ! Correction Arrays
! To save cal sets into the instrument requires knowledge of what
! cal kit and cal types are available. This is a difficult problem
! given that the calibration may be redefined by the user.
! ! Therefore, it's not implemented in this version of the driver.
! NOTIFY "Sorry, storing calibration arrays is not supported.";
! OUTPUT STRING "HOLD;CORROFF;FORM3;";
CASE (Active_Trace) Calc_01;
OUTPUT STRING "FORM3; INPUCALC01;";
CASE (Active_Trace) Calc_02;
OUTPUT STRING "FORM3; INPUCALC02;";
CASE (Active_Trace) Calc_03;
OUTPUT STRING "FORM3; INPUCALC03;";
CASE (Active_Trace) Calc_04;
OUTPUT STRING "FORM3; INPUCALC04;";
CASE (Active_Trace) Calc_05;
OUTPUT STRING "FORM3; INPUCALC05;";
CASE (Active_Trace) Calc_06;
OUTPUT STRING "FORM3; INPUCALC06;";
CASE (Active_Trace) Calc_07;
OUTPUT STRING "FORM3; INPUCALC07;";
CASE (Active_Trace) Calc_08;
OUTPUT STRING "FORM3; INPUCALC08;";
CASE (Active_Trace) Calc_09;
OUTPUT STRING "FORM3; INPUCALC09;";
CASE (Active_Trace) Calc_10;
OUTPUT STRING "FORM3; INPUCALC10;";
CASE (Active_Trace) Calc_11;
OUTPUT STRING "FORM3; INPUCALC11;";
CASE (Active_Trace) Calc_12;
OUTPUT STRING "FORM3; INPUCALC12;";
CASE ELSE;
NOTIFY "Bad CASE for Active_Trace in Data_Array";
END SELECT;
! GOSUB Set_sweep_mode;
END SELECT;
END ACTIONS;
GET ACTIONS;
MATSCALE 0 0 Data_Array; ! Zero out the array first.
FETCH (Sweep_mode1) hold; FETCH Sweep_mode1; NE;
IF STACK THEN;
GOSUB Take_sweep;
END IF;
OUTPUT STRING "FORM3;OUTPDATA;";
GOSUB Get_Data_Array;
IF STACK THEN;
GOSUB Take_sweep;
END IF;
OUTPUT STRING "FORM3;OUTPFORM;";
IF STACK THEN; ! Need to turn correction ON to read raw data arrays
NOTIFY "Turn Correction ON To Read Raw Data Arrays.";
ELSE;
GOSUB Take_Sweep;
OUTPUT STRING "FORM3;OUTPRAW1;";
END IF;
IF STACK THEN;
ELSE;
GOSUB Take_Sweep;
OUTPUT STRING "FORM3;OUTPRAW2;";
END IF;
IF STACK THEN;
ELSE;
GOSUB Take_Sweep;
OUTPUT STRING "FORM3;OUTPRAW3;";
END IF;
IF STACK THEN;
ELSE;
GOSUB Take_Sweep;
OUTPUT STRING "FORM3;OUTPRAW4;";
END IF;
OUTPUT STRING "DISPMEMO;FORM3;OUTPMEMO;";
GOSUB Set_Display_Type; ! Reset the display back where it was !check
CASE ELSE; ! Must be Correction Arrays
IF STACK THEN; ! Need to turn correction ON to read correction arrays
NOTIFY "Turn Correction ON To Read Correction Arrays.";
ELSE;
GET Correction_type;
SELECT Correction_type;
CASE None;
NOTIFY "No data to output";
CASE Response;
OUTPUT STRING "HOLD;FORM3;OUTPCALC01;";
CASE ELSE;
END SELECT;
CASE Resp_and_isoln;
OUTPUT STRING "HOLD;FORM3;OUTPCALC02;";
CASE ELSE;
END SELECT;
CASE S11_1port;
OUTPUT STRING "HOLD;FORM3;OUTPCALC03;";
CASE ELSE;
END SELECT;
CASE S22_1port;
CASE ELSE;
END SELECT;
CASE ELSE;
OUTPUT STRING "HOLD;FORM3;OUTPCALC04;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC05;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC06;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC07;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC08;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC09;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC10;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC11;";
OUTPUT STRING "HOLD;FORM3;OUTPCALC12;";
CASE ELSE;
END SELECT;
END SELECT;
! GOSUB Set_sweep_mode; ! Why ??
END IF;
END SELECT;
END ACTIONS;
END COMPONENT; ! Data_Array
ACTIONS Get_Data_Array;
ENTER STACK FORMAT "#,W"; ! Header = ( '#', 'A' )
DROP; ! Remove Header from top of stack
ENTER STACK FORMAT "#,W"; ! Number of bytes to follow
FETCH 16; DIV; DUP; ! Stack contains number of points
STORE Number_of_points; ! check no_of_data_points
FETCH 1; FETCH 1602; STORE Data_array; ! Store number of points in last element
! for easy arrat-cutting in VEE
FETCH Number_of_points; DUP;
SELECT Channel;
CASE Channel1;
STORE Number_of_points1;
CASE Channel2;
STORE Number_of_points2;
END SELECT;