Most basic bench function generators have only one or two output channels. In some cases, you may need more than two channels. Due to this need, many test engineers are forced to buy very expensive multichannel waveform generators. In this blog, we will look at how to easily time-synchronize multiple basic function generators together without the hassle of using external synchronizing trigger boxes or tedious programming. This mitigates the need to purchase expensive multichannel waveform generators.


There are many applications that require multichannel, time-synchronized waveforms. For example, some devices calibrate with synchronized pulses of varying pulse widths. You can easily simulate these signals using multiple basic function generators. Other applications could include simulating optocoupler decoder signals, multichannel Pulse Width Modulation (PWM) motor controllers, and so on.


How to synchronize two function generators

Today, we’re going to look at how to display four time-synchronized output waveforms using two 2-channel function generators (Keysight 33612A). For our test, we want to generate four 3 kHz pulse trains at 1Vpp. And we want the pulse widths of the four channels to start off at 100 µs and increase by 10 µs for each channel. So, the fourth channel pulse width will be 130 µs.


The first thing to do is wire-configure the rear of the two function generators, as shown in Figure 1. Make one of the function generator’s 10 MHz frequency timebase be the reference for the other function generator. Connect the 10 MHz clock out of the reference function generator to the 10 MHz clock in of the other function generator. (Note that you can fan out the reference clock up to four Trueform function generators).

 

Figure 1

Figure 1. Rear wire configurations on timebase and external trigger synchronization between two function generators.

 

Next thing to do is to connect the "External Trigger out" of the reference function generator to the "Trigger in" of the other function generator. That’s all that is needed for external wire configuration.


User interface system and waveform settings

Reference timebase oscillator and trigger setup

The next step is to set the second function generator’s reference oscillator to external. This ensures that your function generators both operate off from the same timebase.

Button presses: Select menu button, "System" > "Instrument Setup" > "10 MHz Ref Osc" > "Source" > "External".

 

Once that is done, you will see a green "Ext Lock" sign on the top right-hand side of your function generator, as shown in Figure 2.

 

Fig 2 function generator

Figure 2. Second function generator’s reference oscillator and external trigger setup.

 

After setting up your reference oscillator, you will have to set up your trigger for both channels of your second function generator. This tells the second function generator to start at the same time as the first function generator.

Button presses: Select Channel 1 output button and then press the "Trigger" button. Select "Trigger Setup" > "Source" > "Ext" (External). On the same screen menu level, select zero seconds delay and trigger slope on the rising ↑ edge. See Figure 2 on the external trigger menu.

 

On the first reference function generator, just set up the trigger for both channels to manual and zero delays.

Button presses: Set Channel 1 as the source of your trigger out and on the rising ↑edge. Select "Trigger" on Channel 1 > "Trig Out Setup" > "Source" > "CH1". On the same menu level, select the trigger level and trigger out ↑edge.

 

Set up your pulse waveform signals for all your 4 channels (2 channels per instrument)

As mentioned earlier, we want to set up four 1 Vpp pulse trains that are 3 kHz in frequency, and the pulse widths should start at 100 µs and increase by 10 µs increments for every channel. (See Table 1 and Figure 3.)

 

Function Generator12
Output Channel1212
Waveform TypePulsePulsePulsePulse
Frequency3 kHz3 kHz3 kHz3 kHz
Amplitude1 Vpp1 Vpp1 Vpp1 Vpp
Offset0.00 V0.00 V0.00 V0.00 V
Phase0.00°0.00°0.00°0.00°
Pulse Width100 µs110 µs120 µs130 µs

Table 1. Pulse waveform signals set up for 4 channels.

 

Figure 3 function generator

Figure 3. Example pulse setup menu for all 4 channels.

 

Set up burst mode to time synchronize all 4 channels

At this point, if you look at all 4 channels on an oscilloscope, you will see the 4 pulse trains, but they are not time synchronized (Figure 4). All channels run independently, hence they are not initially synchronized.

 

Figure 4
Figure 4. Multichannel outputs that are not time synchronized.

 

To time synchronize all 4 channels, use the Burst mode (Figure 5).

Button presses: Go to all 4 channels, Select "Burst" > "N Cycle" > "# Cycle" > "Infinite".

 

Figure 5 function generator

Figure 5. Burst mode setup menu on function generator.

 

Turn on Burst for all 4 channels. You will be prompted by a blinking light on the "Trigger" button of your Channel 1 on your first function generator. Press the button twice, and instantly, you will see 4 channels of time-synchronized pulse trains (Figure 6).

 

Figure 6
Figure 6. Time-synchronized, multichannel outputs.

 

You will also notice that each pulse width is 10 µs longer than the previous channel. You have just walked through the hardware and front panel configurations to time synchronize 4 channels. Since you can fan out the 10 MHz reference clock to up to four Keysight Trueform function generators, you can have up to 8 time-synchronized channels.

 

Knowing how to set up time-synchronized multiple waveforms does not mean that you cannot be creative with your waveforms. Figure 7 gives you an example of simulating a 3-phase AC generator with 3 sinusoidal waveforms, 120° apart from one another.

 

Figure 7
Figure 7. Simulated 3-phase AC generator 120° apart with multiple function generators.

 

To learn more about phase, frequency, and amplitude signal coupling, please read the Effortlessly Couple or Synchronize Two Signals on a Waveform Generator application brief. I invite you to like and share if you have found this blog helpful. Thanks for reading.

Have you ever set your power supply output voltage to a value and found the voltage at your load was lower than you expected? Many of us have experienced that outcome, and that’s because remote sensing needs to be part of the setup. In this article we are going to share with you three things you need to know about remote sensing to help you get a value you can trust.

 

1. Use remote sensing to regulate voltage at your load

Remote sensing is a feature on many power supplies that allows the power supply to remotely regulate the voltage right at your load. This is accomplished by using a set of remote sense leads that are in addition to your load leads. The power supply uses the voltage on the remote sense lead terminals to sense the voltage right at the load terminals and regulate the voltage right at the load by adjusting the output terminal voltage.


Figure 1 shows the power supply setup using remote sensing. The remote sense terminals are connected to the load at the points where you want the 5 V setting to be regulated. In this case, the power supply regulates 5 V at the load by adjusting its output voltage to 5.3 V to make up for the drops in the load leads. It does this by using the voltage across the sense leads as part of the feedback loop inside the power supply to adjust the voltage on the output terminals.
The purpose of the power supply is to keep the sense lead voltage constant at the setting; the power supply changes the output terminal voltage based on the sense terminal voltage. The input impedance of the sense terminals is high enough to prevent any significant current flow into the sense terminals – this makes any voltage drop on the sense leads themselves negligible.

 

remote sensing compensates load lead voltage drop
Figure 1: Using remote sense to compensate for load lead voltage drop.


2. Use sense leads for overvoltage protection (OVP)

One of our military customers providing DC power to a very expensive device during test asked about the availability of a special option on one of our power supplies. They wanted the option that changed the location of the overvoltage protection (OVP) sensing terminals from the output terminals of the power supply to the sense terminals of the power supply. Since the device under test (DUT) is located quite a distance away from the power supply, they are using remote sensing to regulate the power supply voltage right at the device under test. And since the DUT is very expensive and sensitive to excessive voltage, it’s important to protect the input of the DUT from excessive voltage as measured right at the DUT input terminals.

 

The power supply used, Keysight N6752A installed in an N6700C mainframe, normally uses the output terminals as the sensing location for the overvoltage protection. OVP is used to prevent excessive voltage from being applied to sensitive devices. If the voltage at the output terminals exceeds the OVP setting, the output of the power supply shuts down.

 

Since this customer is very interested in preventing excessive voltage from being applied to the expensive DUT, sensing for an overvoltage condition right at the DUT is important. For the N6752A, Keysight offers a special option (J01) that adds the ability to perform OVP sensing with the sense leads. See Figure 2. with the J01 option added to the N6752A, the customer’s DUT is protected against excessive voltage.

 

OVP sensing at DUT sense lead terminals
Figure 2: OVP sensing is done right at the DUT using sense lead terminals in addition to the output terminals.

 

You may be wondering why the standard OVP would sense at the output terminals instead of at the sense terminals. Probably the biggest reason for sensing at the output terminals is because that approach provides more reliable protection than sensing at the sense leads even though it is less accurate. The output terminals are the power-producing terminals.

 

If the sense leads become inadvertently shorted, the voltage at the output terminals would rise uncontrolled beyond the maximum rated output of the power supply. This uncontrolled high voltage could easily damage any device connected to the power supply’s output leads. So, sensing for an overvoltage condition at the output terminals makes sense. It may not be the most accurate way to protect the DUT, but it is the most reliable given all of the things that can go wrong, such as a wiring error or an internal fault in the power supply.


3. Remote sensing can affect load regulation performance

The voltage load effect specification tells you the maximum amount you can expect the output voltage to change when you change the load current. In addition to the voltage load effect specification, some power supplies have an additional statement in the remote sensing capabilities section about changes to the voltage load effect spec when using remote sensing. These changes are sometimes referred to as load regulation degradation.

 

For example, the Keysight 6642A power supply (20 V, 10 A, 200 W) has a voltage load regulation specification of 2 mV. This means that for any load current change between 0 A and 10 A, the output voltage will change by no more than 2 mV. Also included in the 6642A remote sensing capability spec is a statement about load regulation. It says that for each 1-volt change in the + output lead, you must add 3 mV to the load regulation spec. For example, if you were remote sensing and you had 0.1 ohms of resistance in your + output load lead (this could be due to the total resistance of the wire, connectors, and any relays you may have in series with the + output terminal) and you were running 10 A through the 0.1 ohms, you would have a voltage drop of 10 A x 0.1 ohms = 1 V on the + output lead. This would add 3 mV to the load regulation spec of 2 mV for a total of 5 mV.

 

When you are choosing a power supply, if you want the output voltage to be well regulated at your load, be sure to consider all the specifications that will affect the voltage. Be aware that as your load current changes, the voltage can change as described by the load effect spec. Additionally, if you use remote sensing, the load effect could be more pronounced as described in the remote sensing capability section (or elsewhere). Be sure to choose a power supply that is fully specified so you are not surprised by these effects when they occur.

 

Summary

Remote sense is used to regulate the set voltage at the DUT, compensating for any loss in your leads. Using remote sense will have an impact on regulation performance, which should be considered along with the benefit of compensating for the voltage drop in your leads. Overvoltage protection at the power supply outputs should be used in conjunction with remote sense to protect the DUT.


You can learn more how to protect your DUT against power-related damage by downloading the Protect Against Power-related DUT Damage application note from Keysight.com

Have you ever encountered a scenario in which an AC voltage signal is measured on an electrical circuit that has been completely disconnected? Isn’t it confusing when voltage is measured in the dummy circuits?

 

Stray voltage, sometimes referred to as ghost voltage, is a voltage that appears in an electrical conductor such as a wire, even though the wire is disconnected from an electrical circuit. You may spend hours troubleshooting this circuit and end up realizing that it’s a stray voltage, even though all wires are disconnected!

 

Where do stray voltages come from?

It is very common for electricians and technicians to pull extra wire when facilities or buildings are built and wired. This is just like renovating your house - you will pull extra wire from the conduit for future usage. Normally, these wires are left unconnected. These are the areas where phantom voltage will appear in the circuits.

Wires left unconnected are most likely to be the areas where stray voltage will appear in your circuits. 

 

Why do stray voltages appear?

Stray voltage readings can be caused by capacitive coupling of energized conductors with nearby unused wire. This capacitance increases as the length of the conductor increases. The longer the wire, the more prevalent a stray voltage.

 

Current in an active circuit can also trigger a stray voltage reading; the higher the current in the active circuit, the higher the stray voltage. Stray voltage readings caused by active circuits can range from a few volts up to the voltage of the adjacent conductors. It should be noted that according to Underwriters Laboratories Inc. (UL), stray voltage is not real voltage and it cannot cause physical harm to a person. This is because, even though the voltages may be high, the amount of energy stored in the capacitive coupling is very low.

 

UL also states that care must be taken to ensure that the voltage reading is a stray voltage and not a result of a cable defect or improper installation; as such a situation may result in a shock hazard.

 

Here is an example that illustrates the overall situation. Imagine that you are installing low voltage lighting in a warehouse office, as shown in Figure 1. The warehouse is equipped with two wires running in parallel to the conduit. One is for light A, which is ON, and the other pair of wires will be used to install a new light using a new expansion cable that runs parallel with light A.

 

installation in building or facilityFigure 1: Installation of low voltage lighting in a warehouse office.

 

Before beginning the installation, you check the voltage on the wire using a normal handheld multimeter with high input impedance, and the measurement result shows as 40 volts even though the line is disconnected from the main switch. Now, you suspect that touching conductors has formed a short circuit, causing voltage to leak through the conductor’s insulation. You spend a lot of time troubleshooting and investigating. However, after a thorough investigation, you find that there is no short circuit to ground! The 40 volts displayed on the measurement reading is a phantom voltage reading formed by the unused wire. After all the hard work troubleshooting, you realize you have lost a lot of time troubleshooting a stray voltage.


From this example, we can conclude using a normal handheld digital multimeter to measure such circuit can make it difficult for you to differentiate ghost voltage reading from legitimate readings. Most handheld digital multimeters have high input impedance compared to the impedance of the circuit being measured. The handheld multimeters with high input impedance that is greater than 1 MΩ are designed to place very little load on the circuit under test. In this capacitive coupling situation, a phantom voltage reading is measured by this high input impedance multimeter.

 

In our example, if a low input impedance multimeter had been used to perform the AC voltage measurement, the electrician would have found virtually zero stray voltage. This is because stray voltage is a physical phenomenon involving very small values of capacitance; it cannot energize a load. Using a multimeter with low input impedance will short out the capacitive coupling effect, while using a high input impedance multimeter will not.


The solution

Certain models of Keysight’s handheld multimeters, for example, the U1242C, have a unique feature: a ZLow function (Figure 2) that allows you to switch from high input impedance mode to low input impedance mode to check for the presence of stray voltages. This solution eliminates the need to carry both a low impedance meter and a high impedance meter.

ZLOW function on U1242C handheld DMMFigure 2. U1242C handheld DMM with ZLow function

 

The ZLow function acts like a backup voltage indicator and eliminates the need to carry additional tools for troubleshooting. If a real voltage is measured using the ZLow function, the positive temperature coefficient (PTC) thermistor that is designed as an over current protection will ensure the multimeter always operates in high input impedance.


Use a multimeter with flexible input impedance

Now you know how to detect stray voltages efficiently and effectively using a handheld multimeter with low impedance mode. Keysight offers different handheld multimeters that come with ZLow function that can remove stray voltages from your measurements by dissipating the coupling voltage. Use ZLow to reduce the possibility of false readings in areas where the presence of stray voltages is suspected.


To learn more, download the Stray Voltage Testing Made Easy with U1272A application note.
Check out Keysight.com for more info about Keysight’s handheld digital multimeters.

 

Keysight handheld multimeters

Aside from wireless and fiber optic transmissions, in this modern digital and mixed-signal age, many of our data transmissions go through good old-fashioned conducting cables. The two most common types of cables are differential and single-ended. There are obviously pros and cons in choosing either type of cable, but differentials have many advantages over single-ended cables. Differential signaling is usually used in conjunction with tightly twisted pair wires to reduce or cancel out the generation of electromagnetic noise. Hence, it has superior signal-to-noise ratio and fewer timing errors.

 

There is a Low-Level Differential Signaling (LVDS) standard for electrical transmission and communication protocols that are used in very low voltage and sometimes high-speed data transmissions such as video, graphics, and digital data bus transfers. Other applications include transmitting sensitive analog signals with audio microphones and medical heart monitoring devices.

 

What is a Differential Signal, and How Do I create a Differential Signal?

 

Figure 1 below shows two complementary signals (one signal inverted from the other) being used to create a differential output transmission signal. Here’s how to create differential signals for testing.

 

Differential output derived from two complementary signals (out+/out-).

Figure 1. Differential output derived from two complementary signals (out+/out-).

 

Step 1: Use a function generator to create the first of your two complementary signals (out+/out-). For example, create the (out+) signal on Channel 1.

 

Step 2: Select Dual Channel Inverted Tracking mode (see Figure 2).

 

As a result, Channel 2 will output an identical, but inverted, signal of Channel 1 (out-). Both signals will be amplitude and phase synchronized.

 

Differential channel output function generator setup on a Keysight Trueform function generator.

Figure 2. Differential channel output function generator setup on a Keysight Trueform function generator.

 

Step 3: Combine the differential outputs. To physically combine the two channels into a single differential output, connect the two common connections (the connector shells) of both channels together. Use the middle signal pin of Channel 1 as the high signal path of the differential signal, and use the signal pin of Channel 2 as the inverse return path.

 

Then connect a twisted pair cable to signal pins of Channel 1 and 2, as shown in Figures 3 and 4.

 

Differential signal block diagram.

Figure 3. Differential signal block diagram.

 

Function generator setup based on block diagram.

Figure 4. Function generator setup based on block diagram.

 

On the other end of the twisted pair cable, connect the signal pin of Channel 1 to the signal pin of the receiver’s BNC connector (Differential Signal Input) and the signal pin of Channel 2 to the ground connector shell of the receiver’s BNC (Figure 4).

 

Why and When Do You Use Differential Cables?

 

Basically, a product or systems designer will choose differential over single-ended signaling if:

  • The designer needs to reduce EMI or electromagnetic interference
  • The designer needs to reduce crosstalk or interference from nearby cables
  • The designer needs to transfer very low voltage signals, especially in millivolt range, to simulate bio-signals for medical applications. Low voltage signals are susceptible to noise interference
  • The designer needs to transfer low digital voltage signals to save power
  • The designer needs precise timings of digital signal crossover or digital switching

 

What Type of Function Generator Should You Use to Simulate Differential Signals?

 

Create a Balanced Pair of Signals

Use a function generator with “inverted tracking mode” to create an inverted mirror image of a signal. The signals on both channels will then be perfectly balanced pairs with synchronized amplitude, offset, and phase. Make sure the lengths of your differential twisted pair wires are matched too.

 

Reproduce Actual Signals from Your Design

Consider using an arbitrary function generator and its ability to reproduce actual signals from your design. You can quickly recreate your design signal by capturing it with a modern oscilloscope, then saving the captured trace to a .csv file.

At this point, you can use a USB thumb drive to import the data in the Waveforms > Arbs menu of your function generator to recreate and playback the signal from your design. You can even change your signal’s frequency, amplitude, or offset.

After recreating your signal, you can use the Dual Channel Inverted Tracking mode to recreate the output as a differential signal.

 

Create Arbitrary Waveform Signals

Differential signals are not limited to basic function generator signals such as periodic pulse, sine, square, and ramp signals. A lot of applications need complex waveforms, such as ECG bio-signals for medical applications, automotive CAN bus test signal simulations, telecommunication network test signals, and more.

Keysight’s Trueform function generators provide software tools such as BenchVue Waveform Builder (see Figure 5) and a built-in waveform editor to create your arbitrary waveforms. You can also use tools such as Excel or Matlab to create your arbitrary waveforms and transfer them into your function generator via .csv file format.

 

BenchVue function generator app to create arbitrary waveforms.

Figure 5. BenchVue function generator app to create arbitrary waveforms.

 

Use a Function Generator with the Lowest Jitter

If timing of your signal is very critical, differential signaling is a better option than single-ended signaling since it eliminates the uncertainties of transition crossover points. This takes care of jitter noise from cable transmission.

How about jitter noise from the source of the signal? Not all function generators are built the same! Keysight’s Trueform function generators have superior jitter noise compared to the conventional DDS technology used by most function generators in the market. (See Figure 6.)

 

Trueform technology, shown on the left, has significantly better jitter performance compared to conventional DDS technology.

Figure 6. Trueform technology, shown on the left, has significantly better jitter performance compared to conventional DDS technology.

 

Add Noise to Your Signals

Sometimes, you need an ideal signal with low noise, low distortion, and high signal integrity for your tests. However, at the same time, you want to introduce a controlled imperfect signal with noise for your tests.

Make sure to use a function generator that allows you to add noise to your signals. Press the Modulate button, select Source as Internal/Shape as Noise, and turn on Modulation. (See Figure 7.)

 

Adding variable bandwidth noise to signal for testing.

Figure 7. Adding variable bandwidth noise to signal for testing.

 

Conclusion

 

You can see that it’s easy to create differential signals if you have the right function generator. Choosing a function generator that has built-in differential capabilities will streamline your testing and ensure your signal’s phase and amplitude stay balanced.

 

To get better at using your function generator in the lab, download the Creating Differential Signals with a Waveform Generator application brief.

 

Visit Keysight.com for more information on Keysight’s Trueform function generators.

 

As the world continues to trend toward increased energy savings and green energy sources, more and more heavy machinery and vehicles are becoming electrified. Mechanical combustion engines are being replaced by electric motors as part of this technology trend. As these demands accelerate, higher expectations for reliable and safe power are spurring engineers to put all their brain power into coming up with the most efficient product designs. The last thing that R&D engineers want is to mess with the power supply reliability and create a potential safety issue.

 

Over Voltage and Over Current Protection

 

Today’s system DC power supplies incorporate a variety of features to protect both the device under test (DUT) as well as the power supply itself from damage due to a fault condition or setting mishap. Over voltage protect (OVP) and over current protect (OCP) are two core protection features that are found on most system DC power supplies to help protect against power-related damage. But is that all that you need to know? In this blog, we will also discuss power protection on your devices that do not operate on fixed voltage and current levels.

Over voltage protect helps ensure the DUT is protected against power-related damage in the event the voltage rises above an acceptable range of operation. As over voltage damage is almost instantaneous, the OVP level is set at reasonable margin below this level to be effective; yet it is set suitably higher than the maximum expected DUT operating voltage so transient voltages do not cause false tripping. Causes of over voltage conditions are often external to the DUT.


Over current protect helps ensure the DUT is protected against power-related damage in the event it fails in some fashion, causing excess current, such as an internal short or some other type of failure. The DUT can also draw excess current by consuming excess power due to overloading or from an internal problem that causes inefficient operation and excessive internal power dissipation.

OVP and OCP are depicted in Figure 1 below in an example DUT that operates at a set voltage level of about 48V and uses about 450W of power. In this case the OVP and OCP levels are set at around 10% higher to safeguard the DUT.

 

 

OVP and OCP settings to safeguard an example DUT.

Figure 1. OVP and OCP settings to safeguard an example DUT.

 

Over Power Protection


However, not all DUTs operate over a limited range, as depicted in Figure 1. Consider, for example, that many (if not most) DC-to-DC converters operate over a wide voltage range while using relatively constant power. Similarly, many devices incorporate DC-to-DC converters to give them an extended range of input voltage operation. To illustrate with an example (see Figure 2), consider a DC-to-DC converter that operates from 24 to 48 volts and runs at 225 W. DC-to-DC converters operate very efficiently, so they dissipate a small amount of power and the rest is transferred to the load. If there is a problem with the DC-to-DC converter that causes it to run inefficiently, it could be quickly damaged due to overheating. While the fixed OCP level depicted here will also adequately protect it for over power at 24 volts, you can see that it does not work well to protect the DUT for over power at higher voltage levels.

 

Example DC-to-DC converter input V and I operating range.

Figure 2. Example DC-to-DC converter input V and I operating range.


A preferable alternative would be to have an over power protection limit, as depicted in Figure 3. This would provide an adequate safeguard regardless of the input voltage setting.

 

Example DC-to-DC converter input V and I operating range with over power protect.

Figure 3. Example DC-to-DC converter input V and I operating range with over power protect.


Since an over power level setting is not a feature that is commonly found in system DC power supplies, this would then mean having to change the OCP level for each voltage setting change, which may not be convenient, desirable, or in some cases, practical to do. However, in the Keysight N6900A and N7900A advance power system DC power supplies, it is possible to continually sense the output power level in the configurable smart triggering system. This can then be used to create a logical expression to use the output power level to trigger an output protect shutdown. The N7906A software utility was used to graphically configure this logical expression, and then it was downloaded it into the advance power system DC power supply, as shown in Figure 4. Since the smart triggering system operates at hardware speeds within the instrument, it is fast-responding, an important consideration for implementing protection mechanisms.

 

N7906A software utility graphically configuring an over power protect shutdown.

Figure 4. N7906A software utility graphically configuring an over power protect shutdown.

 

A glitch delay was also added to prevent false triggers due to temporary peaks of power being drawn by the DUT during transient events. While the output power level is being used here to trigger a fault shutdown, it could just as easily be used to trigger a variety of other actions.

 

Summary

 

We have discussed that advance system power supplies can provide over voltage and over current protection as well as protection for over power conditions. For more information on protection against power related damage, download the Protect Against Power-related DUT Damage During Test application note.

Electronics are often designed to work anywhere in the world with local power. An AC source can create the various voltage-frequency combinations used in different countries and even measure the current and power consumption. 

Figure 1 Power supply that accepts 100 Vac to 240 Vac

 

 

 

 

 

 

 

 

 

 

Figure 1. A power supply that accepts 100 Vac to 240 Vac.


You can automate the process using Visual Basic for Excel to program an AC6801B AC source. Excel lets you capture and share results in a neatly organized table. In this example, we chose eleven countries to demonstrate the wide variety of voltages and frequencies used around the world. Some countries use multiple voltages and plug types, and those displayed are selected to illustrate a broad range. The table is easily modifiable, and you can add more countries by increasing the rows as well as the country count in cell D6.

 

Figure 2 table with various voltages and frequencies

Figure 2. An example table of various voltages and frequencies. Row number and column letter shown in red.

 

The Visual Basic program reads the voltages and frequencies from the table and sets up the AC source to output them. The program pauses for the number of seconds provided in column F, allowing the test device to settle. The AC source then makes several measurements and adds them to each row of the table.

 

Figure 3 AC source adds measurements to each row in table

Figure 3. The AC source output is set to 100 Vac at 50 Hz. Three measurements are added to the table after the delay. The process is repeated for each country.

 

Getting started with Visual Basic for Excel

If you are not currently using Visual Basic for Excel, you need to display the developer tab. It is included in Excel but is not shown by default. You can find instructions online for adding the tab to the version of Excel you use. The second step is adding Keysight IO Libraries Suite. In this example, all the commands are sent to the AC source using the Keysight IO library. With the software installed, open the developer tab and create the global variables.

 

Figure 4 Opening VBA project

Figure 4. Opening the VBA project and creating the global variables.

 

Opening a connection to the AC6801B

The following subroutine creates a connection to communicate with an instrument based on its VISA address. The VISA address is read from the table, making it easy to update. You can use the Keysight Connection Expert, which is installed with the IO Suite to read the instrument address. Watch this YouTube video to see the steps to locate the instrument VISA address and verify the connection.

 

Figure 5 Subroutine to open a connection to instrument

Figure 5. Subroutine to open a connection to the instrument.

 

You can insert a button onto the spreadsheet (Developer > Insert > Form Control > Button). You are then prompted to add the name of the subroutine that will run when the button is clicked.

 

Figure 6 Excel with controls and LXI address

Figure 6. The spreadsheet with the first two controls and the LXI address.

 

Resetting and configuring the instrument

The subroutine initConfig_Click() has three primary roles: to clear the AC source, to configure DUT protection, and to select the upper range and turn on the AC source with a known voltage. Once the instrument is reset, it is set to output only AC voltage, and the voltage is limited by a min and max value. Obviously, overvoltage can cause damage to a device, but too low a voltage is just as harmful. In the case of a power supply, too low a voltage leads to higher current to meet the power needs of the load. In addition, a current limit is set to protect the DUT from damage due to excessive current. The time required to run the test is reduced by using a single range. The AC6801B is a dual range source and provides additional current in its low range. For our device, the upper range has enough current to test the device. Lastly, a voltage-frequency combination is selected, and the output is turned on. The primary goal of this program is to demonstrate some of the AC source’s capability. With the output turned on, the AC source displays measurements on the front panel.

 

Figure 7 Subroutine to put AC source in known state
Figure 7. Subroutine to put the AC source in a known state.

 

Step through the list of countries

The createSequence_Click() subroutine populates the table with the power measurements for each country. It uses the number specified in cell D6 to determine the number of loops necessary to complete the table. In each loop, the voltage-frequency is pulled from the table, and the source is set to output the combination. The power measurements are made after the delay in column F, which allows the device to adjust to the new voltage and frequency. Three power measurements were selected from a choice of 17 different measurements. As each measurement is made, the results are added to the table.

 

Figure 8 Subroutine to apply voltage-frequencies from table
Figure 8. Subroutine to apply the voltage-frequencies from the table and populate the measurements.

 

Closing the instrument connection

Closing the instrument connection releases the resources. Once the connection closes, an error is generated if you attempt to send commands to the instrument.

 

Figure 9 Closing instrument connection
Figure 9. Closing the instrument connection.

 

Reading instrument errors

The read error subroutine is useful while creating or modifying a program. It pulls the error strings from an instrument one string at a time. You may need to run it multiple times to clear all the errors. It is designed to run independently of the rest of the program. It assigns resources, opens the IO, and then closes the IO.

 

Figure 10 Subroutine to query error strings
Figure 10. A subroutine to query the error strings.


Summary

The universal adapter test program is designed to demonstrate some of the AC6800 Series capabilities and a method to document the measurements using Excel with Visual Basic. Adding a main program to call each of the subroutines simplifies the program because it’s not necessary to connect each routine to a separate control button. The program ran many times over a two-week period using the AC6801B, and the results are repeatable. Often the results are identical when formatted to show a single decimal place. The program is easily modifiable to make additional measurements. A couple of real-world applications are to characterize a group of power supplies and statically determine the max power usage or verify results from a remote facility. Using a second AC source, it is easy to reproduce results with a production facility located on the other side of the globe.

A search of the internet provides articles and videos showing the general operation of a synchronous motor. I thought it would be interesting to make some actual measurements and display some real-world voltage and current waveforms. I was fortunate to find a synchronous motor that had external wiring to its two coils, allowing independent characterization of each coil. The motor is small, but the characterization would scale to a larger motor.

 

Synchronous Motor Basics

An AC signal powers a synchronous motor. It uses the positive/negative power cycle to create a changing magnetic field. Our motor uses two coils that are 90 degrees out of phase. The fields change such that the magnetic rotor turns to keep aligned with the fields. Synchronous 3-phase motors are popular due to the simplicity of using each phase to create a rotating field. You can view many good animations of how a synchronous motor works on the internet.

 

Drawing of synchronous motor with two coils

Figure 1. A drawing of a synchronous motor with two coils. The colors represent the colors of the motor leads.


The AC source causes the motor to rotate at a rate proportional to its frequency (typically measured in Hz). Our example motor is powered by a 60 Hz AC signal and rotates at 3600 RPMs. The synchronous motor can turn in either direction. It starts turning toward the attracting magnetic force, which could be either of the coils depending on its configuration. A minimum voltage is required to create a magnetic field strong enough to cause the rotor to spin. Once spinning, the AC source voltage does not affect the speed of rotation.

 

Changing location of AC source
Figure 2. Changing the location of the AC source to cause a different direction of rotation.

 

Electrical Connections to the Motor

I wanted to characterize the voltage and current through each coil. I used our PA2201A power analyzer to characterize each coil. Each channel of the power analyzer has an input for voltage and an input for current. It can make  measurements of each input or measurements that combine two inputs. While not shown in Figure 2, I added a switch to connect the AC source to coil A or coil B.

 

Drawing of motor with power analyzer connected
Figure 3. An electrical drawing of the motor with a power analyzer connected. The power analyzer connections are color-coded. Channel 1 is yellow and Channel 2 is green.

 

The AC6801B AC source supplies the 110 VAC input power. The AC source will characterize all the power parameters for the motor. We can also vary the frequency of the input voltage.


Real-World Measurements

We will start with the AC source connected to the coil A as well as channel 1 of the power analyzer, shown in Figure 3. We expect to see the 110 V on channel 1 (top yellow waveform) of the power analyzer. Notice the coil B voltage shown on channel 2 (top green waveform) lags channel 1 almost perfectly by 90 degrees.

 

Screen capture of PA2201A power analyzer
Figure 4. Screen capture from the PA2201A. The display is split into three parts: the top is voltage, the middle is current, and the bottom is power. Yellow waveforms represent channel 1, coil A, and green waveforms represent channel 2, coil B.


I do not work with inductors and capacitors every day and was surprised that coil B receives much more current and power. If you look at the equations for reactance, it makes sense. The capacitor reduces the reactance X in the branch with coil B, X = XL - XC. More current flows through this branch, which creates more voltage across coil B.


Reactance through motor
Figure 5. The diagram shows the reactance through each branch of the motor.

 

PA2201A power analyzer display
Figure 6. Measuring the power of channel 1, coil A, and channel 2, coil B, with PA2201A power panels.

 

The AC source is measuring the input to the motor, which is consuming 0.05 ARMS, 5.1 W, and 5.4 VA. Coil B uses most of the energy. I was also surprised to learn the coil's reactance is nearly 50% resistive, 50% inductance. A phase angle ɸ of 45% or power factor PF .707 would be precisely a 50% mix, and the measurements were only slightly more inductive. You can see the phase angle (the delay between voltage and current) in the Figure 6. Lastly, I had expected the current waveforms to be more sinusoidal. In reality, the motor acts as a transformer and sends energy between the coils. If I remove the capacitor, you can see 50 V across coil B; No current flows, as I left the branch open.

 

 PA2201A power analyzer display

Figure 7. With the capacitor removed, the motor can turn in either direction. Energy in coil A creates a voltage on coil B. On the left, the motor is turning in a counterclockwise direction (normal), and on the right it is turning clockwise. Notice in the first case, the coil B voltage (shown in green) lags, and in the second it leads.

 

Reversing the Motor Direction

Applying the source voltage to coil B instead of coil A causes the motor to run in the opposite direction.

 

Motor drawing
Figure 8. Applying the source voltage to coil B, which causes the motor to turn clockwise.

 

PA2201A power analyzer display
Figure 9. Measurements of the two coils with the source connected to coil B.

 

The coils are not perfectly symmetrical, as the power through each branch is a bit lower than in the counterclockwise setup. Using the AC source, the overall power measurements are also lower, reduced to 4.8 and 5.1 VA. You will notice the voltage in coil B, green waveform, leads coil A. You can compare Figure 9 and Figure 6 to see the differences.

 

Changing the Source Frequency

The AC source can vary its frequency from 40 Hz to 500 Hz. Changing the frequency of the source affects the speed of the motor. It also changes the impedance of the circuit, as the reactance of the capacitor and inductors are a function of frequency.

 PA2201A power analyzer display
Figure 10. The waveforms on the left were captured with a source frequency of 40 Hz, and on the right the source frequency was increased to 80 Hz.


Summary

I was impressed with the motor’s overall power factor. The AC source measured the PF to be 0.94. Earlier, we measured the current through each branch of the circuit. The total current consumed by the motor is a point-by-point summation of the two current waveforms (i.e., the sum of the current through each of the coils). A PF of 0.94 indicates that the summation has nearly identical phase to the input voltage. The power analyzer can sum the two current waveforms and display them overtop the input voltage.

 

PA2201A power analyzer display
Figure 11. The pink waveform represents the total current used by the motor. It is a mathematical summation of the current waveforms shown at the bottom of the screen.


As you can see, the system has a PF near 1.0, as the input current and voltage have nearly identical phase.


The capacitor used with our motor is considered a run or permanent split capacitor, as it maintains the 90-degree shift in the coil voltages. People sometimes generically refer to a capacitor used with a motor as a start capacitor. A start capacitor is disconnected by a switch once the motor starts spinning.


To overcome the difficulties in changing the speed of our single-phase motor, the manufacturer added a gearbox so it spins at 33.3 RPMs.


The AC6801B AC source made it easy to supply the required AC power (110 VAC at 60 Hz) while making power measurements. The measurements include current, real power, apparent power, reactive power, and power factor. The PA2001A was used to characterize the voltage and current through each coil.

Using a digital multimeter (DMM) to perform measurements is very common in today’s world. Technicians use multimeters for equipment servicing, engineers use multimeters to troubleshoot, students use multimeters for lab research, and so on. Since digital multimeters have many functions, you need to know how to properly use your multimeter. Most multimeter failures are caused by improper use.

 

In this blog, you will learn four tips to avoid damaging your digital multimeter:

 

  • Read the warning labels and specifications

Before you begin taking measurements with your digital multimeter, you should read the warning labels and specifications. Do not exceed the values provided in the specifications guide or as indicated by the yellow warning labels on the instrument. Always refer to the specification guide for conditions required to meet the listed specification.

 

 Example of the 34470A digital multimeter’s rear panel showing warnings on the maximum voltage input and maximum current input.

Figure 1. Example of the 34470A digital multimeter’s rear panel showing warnings on the maximum voltage input and maximum current input.

 

  •  Ensure proper grounding 

Always use the three-prong AC power cord supplied with the instrument. Proper grounding of the instrument will prevent a build-up of electrostatic charge that may be harmful to the instrument and you. Do not damage the earth-grounding protection by using an extension cable, power cable, or autotransformer without a protective ground conductor. It is good to check the AC power quality and polarity; Typical required AC voltage is 100 V, 120 V, 220 V ± 10%, or 240 V +5%/–10%. Typical expected grounding wire resistance is < 1 Ω; The voltage between neutral and ground line is < 1 V. If needed, you can install uninterruptible power supply [UPS] to power your meter.

 

  • Avoid overpowering the digital multimeter

You can avoid damaging your digital multimeter by anticipating the signal level you’ll measure and presetting the proper signal range on the DMM. Overpowering the digital multimeter can damage the components inside the meter. For example, Figure 2 shows that the maximum voltage input of a 34461A digital multimeter is 1000VDC and 700 VAC. Before turning on or off the connected equipment or the DUT, reduce the signal level to the minimum safety level. This will prevent unexpected voltage or current swell or sag from affecting the input or output of your instrument.

 

Keysight’s 34461A digital multimeter front panel.

Figure 2. Keysight’s 34461A digital multimeter front panel.

 

  • Check for proper temperature and humidity

You need to keep your multimeter in a clean and dry environment. Typical temperature for storage condition is between – 40 and 75 °C; Typical humidity is < 95% RH. The DMM’s optimal operating temperature should be from -5°C to 23° You also need to ensure proper ventilation among racks so the temperature does not go up if all instruments are in use at same time. You should also frequently inspect and clean the cooling vents and fans.

 

Take good care of your instrument!

 

Using your instrument properly helps you and your organization save on maintenance costs. Therefore, it is always good to know the basic tips mentioned above to prevent damage to your multimeter.

 

To learn more, download the Tips for Preventing Damage to Digital Multimeters application note.

Visit Keysight.com or more info about Keysight digital multimeters.

In electronics design and testing, you sometimes want to have two synchronized clock signals that are related by a frequency ratio; One clock needs to maintain a certain frequency ratio with the other clock.

 

Or perhaps you want to simulate an amplifier with an offset; The amplifier output needs to maintain a defined offset from the input amplitude.

 

These requirements may sound basic, but building you own clock reference device or a frequency/amplitude coupling circuit takes time and resources. It’s much easier to use a dual-channel function generator that has built-in signal coupling.

 

Let’s look at how to generate frequency-coupled and amplitude-coupled signals using a dual-channel function generator.

 

How to Generate Two Frequency-coupled Signals with a Function Generator

 

Frequency coupling allows you to specify the frequency relationship of two signals using either a ratio (multiplying) or an offset (adding). Figure 1 shows an example of a frequency coupling ratio of “3.” When the frequency of Channel 1 is set to 3 kHz, the frequency of Channel 2 automatically sets to 9 kHz (Figure 2).

 

Dual-channel frequency coupling setting on ratio on a Keysight 33612A Trueform function generator.

Figure 1. Dual-channel frequency coupling setting on ratio on a Keysight 33612A Trueform function generator.

 

Frequency ratio of 3 as observed on the oscilloscope (see box marked in red).

Figure 2. Frequency ratio of 3, as observed on the oscilloscope (see box marked in red).

 

Here is an example of frequency coupling by offset where the offset difference between Channel 2 and Channel 1 is 2 kHz (Figure 3). When the frequency of Channel 1 is set to 3 kHz, Channel 2 automatically tracks Channel 1 and outputs a 5 kHz signal (see Figure 4).

 

Dual-channel frequency coupling setting on offset.

Figure 3. Dual-channel frequency coupling setting on offset.

 

Frequency offset of 2 kHz as observed on the oscilloscope (see box marked in red).

Figure 4. Frequency offset of 2 kHz as observed on the oscilloscope (see box marked in red).

 

How to Generate two Amplitude-coupled Signals from a Function Generator

 

You can also couple the amplitude and offset of two signals (Figures 5 and 6). Enabling this dual-channel feature and setting the amplitude and offset saves you configuration time. Rather than setting the amplitude and offset of both channels independently, the function generator will keep track of your settings for you. You can even couple signals of different waveform shapes! Figure 6 shows how you can control the amplitude of a square wave and sine wave using only one amplitude setting.

 

Dual-channel amplitude and offset coupling setting on a Keysight 33612A Trueform function generator.

Figure 5. Dual-channel amplitude and offset coupling setting on a Keysight 33612A Trueform function generator.

 

Amplitude and offset coupling of two signals as observed on an oscilloscope.

Figure 6. Amplitude and offset coupling of two signals as observed on an oscilloscope.

 

Benefits of Dual-channel Frequency Coupling and Amplitude Coupling

 

Here are some situations where coupling the amplitude and/or frequency of your test signals is useful.

 

Creating multiple reference pulse clocks to test circuitry

Some electronic devices operate with multiple frequency reference clocks. Dual-channel frequency coupling comes in handy for quick design verifications.

 

Creating two very different arbitrary waveform signals that track each other in frequency

Consider this example of simulating a pacemaker pulse and an ECG arbitrary pulse wave that has the same frequency. When combining these two signals together, pacemaker manufacturers can test a pacemaker’s pulse rejection capability using the resulting signal, shown in Figure 7.

 

Two frequency-coupled signals combined into a single signal output.

Figure 7. Two frequency-coupled signals combined into a single signal output.

 

Testing the differential gain of an amplifier

Testing the differential gain of an amplifier requires synchronized input signals. A function generator and oscilloscope are the ideal test setup for this (Figure 8). If you have perfectly identical input signals for the amplifier, you should see a zero-difference output (straight line), as shown by the oscilloscope’s output in Figure 9. If the output is not flat, you know there’s something wrong with your amplifier.

 

Function generator and oscilloscope set up to test an instrumentation amplifier.

Figure 8. Function generator and oscilloscope set up to test an instrumentation amplifier.

 

Green and blue lines are identical input signals into the Op Amp, and the yellow line is the resultant output from the Op Amp.

Figure 9. Green and blue lines are identical input signals into the Op Amp, and the yellow line is the resultant output from the Op Amp.

 

Using the dual-frequency coupling with ratio of 2, the output of the amplifier will show the differential gain output (yellow line), as shown in Figure 10.

 

Green and blue lines are frequency-coupled signals (with a ratio of 2) generated by a function generator for the input of Op Amp, and the yellow line is the resultant output from the Op Amp.

Figure 10. Green and blue lines are frequency-coupled signals (with a ratio of 2) generated by a function generator for the input of Op Amp, and the yellow line is the resultant output from the Op Amp.

 

Benefits of Frequency Coupling and Amplitude Coupling

 

Clearly, dual-channel tracking simplifies your configuration settings. You don’t have to configure channels separately, which is often tedious and repetitive. As a result, your testing will be less error prone and more efficient. If you are using a programming interface, automatic frequency coupling and amplitude coupling simplifies your code.

 

Now that you know how to couple two signals together, you should be able to perform fast simulation on your products or processes. Thanks to this coupling capability, the signal generation process has never been simpler, quicker, or less frustrating.

 

To learn more about frequency coupling and amplitude coupling your signals using a Keysight function/waveform generator, read this app brief “Effortlessly Couple or Synchronize Two Signals on a Waveform Generator.”

A typical part of an engineer’s job is to perform measurements with a multimeter. For example, if you are working as a building maintenance engineer or electrician, your daily routine may require you to measure power from AC mains or other high voltages. Before you start performing a measurement, what is the first thing that comes to your mind? Should you just grab the nearest handheld multimeter available? No!


Imagine you need to select a helmet. You need to select one with high quality to protect your head. But a helmet’s structure, design, and protective ability vary for different kinds of activities. A helmet designed for rock climbing needs to protect you against small rocks, falling objects, and sharp faces. A bicycle helmet needs to protect your head during impact to reduce the likelihood of injury in the event of an accident. A motorcycle helmet needs to protect you against high-speed impacts with the road and other vehicles. It’s important to choose the right helmet for your activity.


Likewise, different multimeters are designed with different levels of protection against common electrical hazards.

 

For your own safety and the safety of those near you, you must choose a multimeter that is designed and tested to protect you against electrical hazards you might encounter. Remember you only have one life, and there are no second chances. Here are some safety considerations that you need to take note of before you start making measurements with a multimeter.

 

1. Understand your multimeter’s safety indicators

Safety certification is important to ensure the multimeter is compliant with the relevant safety standards. Usually, the manufacturer of the multimeter will obtain safety certifications from third-party independent testing agencies, such as the Canadian Standard Association (CSA), to ensure the product has been tested and complied to the relevant safety standard.

 

Products that successfully pass the independent testing are labeled with the logo of the independent testing agency on the back of the multimeter.

 

The accessories used with a multimeter ─ like probes ─ should also be tested and marked with a third-party safety agency logo. Before you start measuring, be sure to select a multimeter and accessories that have passed these tests.

Some independent testing agency logos are as follows:

VDE certification markTuV technical inspection association certification mark CSA Group certification mark ETL SEMKO certification mark

Be cautious of the CE marking:

CE certification markThe “CE” marking is an abbreviation for "European Conformity" (from the French phrase “Conformité Européene”). The CE marking indicates the product's conformity to the applicable European Union safety, health, and environmental requirements, a mandatory conformity mark on all products sold in the European Union.

 

Manufacturers are permitted to self-certify. They must meet the standards, issue their own Declaration of Conformity, and mark the product “CE.” Therefore, the CE marking is not a guarantee of independent testing.

 

For safety purposes, you should not accept a multimeter that has only a CE mark unless you know the manufacturer to be trustworthy and you have reviewed the manufacturer’s Declaration of Conformity.

 

2. Pay attention to voltage rating of the measurement circuit and measurement limit of your multimeter

Before you start to perform a measurement, you need to understand the maximum voltage rating of your circuit. Use a multimeter that will be able to withstand the maximum voltage of the circuit. In general, manufacturers label the multimeter’s measurement limits on the front panel. A multimeter provides protection circuitry to prevent damage to the instrument and protect against the danger of electric shock, provided the measurement limits are not exceeded.

 

To ensure safe operations of the instrument, do not exceed the measurement limits shown on the front panel of the multimeter.


3. Take precautions when performing live measurements

Here are a few safety precautions you should take when making live measurements with a multimeter:

  • When measuring a live circuit, use insulated tools like safety glasses, insulated mats, and insulated gloves.

insulated gloves with handheld DMM

Figure 1. Wear insulated gloves when using a multimeter.

 

  • Inspect the test leads for damaged insulation or exposed metal. Check the test leads for continuity. If the test leads are damaged, replace them before you use the multimeter.
  • It is recommended that you disconnect circuit power and discharge all high-voltage capacitors before testing resistance, continuity, diodes, or capacitance.
  • When making measurements, always connect the common test lead before you connect the live test lead. When you disconnect the leads, disconnect the live test lead first. Avoid holding the test lead in your hands to minimize personal exposure if transients occur. Before use, verify the multimeter’s operation by measuring a known voltage.


Conclusion – Safety First!

Safety is always the top priority no matter what you are measuring, so there are a few precautions you should take. Always choose a multimeter with a voltage rating higher than the circuit you are measuring. Remember to check that the multimeters and probes you are going to use are marked with third-party safety agency logos such as CSA, ETL, TÜV, or VDE. Do not overlook the safety of the probes! With safety in mind, you will be assured that the high voltage goes into your measurement instrument instead of you!

 

To learn more, download the Think SAFETY when Selecting a Handheld Multimeter application note.

For more info about Keysight’s DMMs, visit www.keysight.com/find/dmm

How can you determine which digital multimeter (DMM) is best for your application?

There are five key specs you need to consider before purchasing a digital multimeter to make sure you’re picking the right DMM for your testing needs: number of display digits, counts, range, resolution, and accuracy. Let’s look at each of these in more detail.

 

1. Display Digits

In general, most manufacturers specify a DMM display by digits of resolution. For example, the digits of resolution for Keysight digital multimeters goes from 3 ½ digits to 8 ½ digits.

 

A 5 ½ digit DMM, for example, has five full digits that display values from 0 to 9, and a fractional digit. Fractional digit is the most significant digit in the display. It is the ratio of the maximum value the digit can attain over the number of possible states. For example, a ½ digit has a maximum value of one and has two possible states (0 or 1). A ¾ digit has a maximum value of 3 with four possible states (0, 1, 2, or 3). The higher the digits of resolution, the more precise and accurate your measurements will be. It also means that a multimeter with higher digits of resolution will be more expensive.


2. Counts

Nowadays, manufacturers have started to specify the display in terms of “count” because “digits of resolution” often creates confusion. The count of a DMM refers to how large a number the multimeter can display before it changes the measurement ranges and how many digits it can show in total. This affects how precise a measurement the DMM can display. For example, a 4½ DMM can also be specified as a 19,999-display count or 20,000 display count multimeter.


3. Resolution

Resolution is defined as the smallest change in an input signal that produces a change in the output signal. Resolution will be improved when the DMM’s measurement range is reduced. Ultimately, you want the multimeter to display the best resolution of your measurement reading. You can play around with the multimeter and select the measurement range that gives you the reading with optimum resolution.

 

4. Range

Range is correlated to resolution, the resolution display on the multimeter will depend on the measurement range that you select.

To choose a digital multimeter, you need to know the minimum and maximum measurement range, and the resolution you require. This information is usually available in the DMM’s data sheet.

 

4.1 Autorange vs. Manual Range Multimeters

Most of the digital multimeters today offer both auto ranging measurements and manual ranging measurements.

 

4.1.1 Auto Ranging Measurements

For auto ranging measurements, all you have to do is select your desired measurement functions and let the multimeter automatically choose the best range. Normally, you should select auto ranging when you do not know the potential measurement reading range.


4.1.2 Manual Ranging Measurements

Manual ranging measurements are normally selected when the desired measurement value is known. Based on your measurement value, you can choose the desired measurement range. If the measurement value is unknown and you would like to use the manual range, it is recommended that you use the “step down” method. Start with the highest range and then step down to lower ranges to achieve the optimum resolution on the display.

 

5. Accuracy

The accuracy of a DMM is different from its display resolution. The accuracy is the maximum allowable limit of error in the readings. Normally, manufacturers display DC voltage (DCV) accuracy as a benchmark when comparing with other manufacturer specifications, as DCV has better accuracy compared to other functions. The accuracy specification is expressed as ±((% of reading + % of range).

 

Summary

There is a range of digital multimeters available on the market. To select the right multimeter, you need to know the target application, the measurement range, and required resolution. If you require a higher measurement accuracy, make sure to select a DMM with a higher resolution.

 

To learn more, download the How to Select a Handheld DMM That is Right for You application note.

For more information about Keysight’s DMMs, visit www.keysight.com/find/dmm.